Error code 4001 for philips BV libra

(1)

Adulf

Error code 4001 for philips BV libra

April 24, 2020 10:15

Dear friends,
The display screen on my philips BV libra C arm doesn't come on and I recently replaced the X2-key panel cable ( SHA1-X2 ) black in colour. Also the monitor on the workstation displays error in startup procedure : 4001
I also have to press and hold the power button for the equipment to come on.
Please kindly help.

Thanks

Log inor Register

to rate and post a comment

(165)
(67)
(2)

Sergey Zheleznyak

re: Error code 4001 for philips BV libra

April 27, 2020 10:37

Hello, probably this information can give you a hint about what could be a problem:
4001 SCSI_E_READ DFI disk failure. CAUTION: After the following actions the PATIENT DATA, DFI
CONFIGURATION DATA and DFI IP SETS are LOST. Start the DFI power-up service menu and low-level format the DFI
disk. Re-install the DFI. See service for more solutions.

1 person liked this
Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: re: Error code 4001 for philips BV libra

May 25, 2020 11:03

Dear all,

The Low-Level format function is only by the DFI 1.
The BV Libra can have a DFI 1 or a DFI 2.
By the DFI 2 you must have the DFI Erase Tool Software.

Kind regards
Willi

2 people liked this
Log inor Register

to rate and post a comment