Dear friends,
The display screen on my philips BV libra C arm doesn't come on and I recently replaced the X2-key panel cable ( SHA1-X2 ) black in colour. Also the monitor on the workstation displays error in startup procedure : 4001
I also have to press and hold the power button for the equipment to come on.
Please kindly help.
Hello, probably this information can give you a hint about what could be a problem:
4001 SCSI_E_READ DFI disk failure. CAUTION: After the following actions the PATIENT DATA, DFI
CONFIGURATION DATA and DFI IP SETS are LOST. Start the DFI power-up service menu and low-level format the DFI
disk. Re-install the DFI. See service for more solutions.
Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty!
Email P/N for pricing: sales@Relaxationinc.com
(P) 813-925-1314, www.relaxationinc.com
Adulf
Error code 4001 for philips BV libra
April 24, 2020 10:15
Dear friends,
The display screen on my philips BV libra C arm doesn't come on and I recently replaced the X2-key panel cable ( SHA1-X2 ) black in colour. Also the monitor on the workstation displays error in startup procedure : 4001
I also have to press and hold the power button for the equipment to come on.
Please kindly help.
Thanks
to rate and post a comment
Sergey Zheleznyak
re: Error code 4001 for philips BV libra
April 27, 2020 10:37
Hello, probably this information can give you a hint about what could be a problem:
4001 SCSI_E_READ DFI disk failure. CAUTION: After the following actions the PATIENT DATA, DFI
CONFIGURATION DATA and DFI IP SETS are LOST. Start the DFI power-up service menu and low-level format the DFI
disk. Re-install the DFI. See service for more solutions.
to rate and post a comment
Herrmann
re: re: Error code 4001 for philips BV libra
May 25, 2020 11:03
Dear all,
The Low-Level format function is only by the DFI 1.
The BV Libra can have a DFI 1 or a DFI 2.
By the DFI 2 you must have the DFI Erase Tool Software.
Kind regards
Willi
to rate and post a comment