Gustavo is correct about error 68 being an acquisition timeout. You have it connected to a Sytec so, it must be an analog interface. I would guess that you probably do not have the proper machine codes stored in the IMS.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
illan ivanovich
Error code 53 Camera 3m851 GE Sytec3000
October 12, 2006 11:23
help me please... i have error 53 in the camera 3m 851...i dont have service manual i need help thanks
to rate and post a comment
Ralph Childs
Re: Error code 53 Camera 3m851 GE Sytec3000
October 13, 2006 09:01
Illan,
Error 53 is a failure in the memory access. It is either due to bad memory partition or the incorrect system set-up between the MCS and IMS.
Scooter
scanner@eim.ae
to rate and post a comment
illan ivanovich
Re: Error code 53 Camera 3m851 GE Sytec3000
October 13, 2006 11:41
ok i moved encoders sw7 and sw8 and now i have the error 68.. thanks for your reply
to rate and post a comment
Gustavo Adolfo Castano
Re: Error code 53 Camera 3m851 GE Sytec3000
October 13, 2006 04:52
error 68 is time out error, when trying to get the image data. You may check the Analog/digital cable or you have a board not working
to rate and post a comment
illan ivanovich
Re: Error code 53 Camera 3m851 GE Sytec3000
October 13, 2006 06:22
these erros apper after I moved the switch (sw1-sw9) I'm very stupid i don't remember the values...help me please...
to rate and post a comment
Ralph Childs
Re: Error code 53 Camera 3m851 GE Sytec3000
October 14, 2006 02:00
Illan,
Gustavo is correct about error 68 being an acquisition timeout. You have it connected to a Sytec so, it must be an analog interface. I would guess that you probably do not have the proper machine codes stored in the IMS.
Scooter
to rate and post a comment