If you have another xli which is functional, you can copy the boot up data from the functional unit onto a floppy diskette and then install into the unit with error 6503. If you do not have another ekg to copy from contact me at stanpoole@bellsouth.net.
Stan Poole
Medical maintenance Consultants
Norcross, Ga.
Online auction of modern high-end laboratory and biopharma research equipment from the recent bankruptcy of Spexis AG in Allschwil near Basel, Switzerland. Start bidding today! closes on June 4. info@dome-auctions.com..+31(0)20 808 2880
Vorick Picou
Error Code 6503 XL1 ECG Machine
November 28, 2011 03:34
Does anyone has the disk for the XL1 ECG machine, I am getting Error Code 6503
to rate and post a comment
Stan Poole
Re: Error Code 6503 XL1 ECG Machine
November 28, 2011 05:54
If you have another xli which is functional, you can copy the boot up data from the functional unit onto a floppy diskette and then install into the unit with error 6503. If you do not have another ekg to copy from contact me at stanpoole@bellsouth.net.
Stan Poole
Medical maintenance Consultants
Norcross, Ga.
to rate and post a comment