Error code E 4512 or E4511

July 17, 2017 04:28

Getting an errors E4511/E4512 on Konica drypro 832. Can someone guide me.

Log inor Register

to rate and post a comment

(2681)
(674)
(21)

Lenny Place

re: Error code E 4512 or E4511

July 18, 2017 07:45

4511 E Film jam has occurred in the sub-scan unit. Film is jammed at the V-sync sen-sor. • Justification unit nip operation failure. • Justification transport motor damaged, or poor connection. • Justification nip home sensor damaged, or poor connection. • V-sync sensor damaged, or poor connection. • Sub-scan motor damaged, or poor connection. • Sub-scan entry/exit nip solenoid damaged, or poor connection. • Check operation and, if damaged, replace the sub-scan unit (including the justification unit). • Check cable bundle connections and operation: if dam-aged, replace. • Check cable bundle connections and operation: if dam-aged, replace. • Check cable bundle connections and operation: if dam-aged, replace. • Check cable bundle operation and connections and, if damaged, replace the sub-scan unit (including the justifi-cation unit). • Check cable bundle operation and connections and, if damaged, replace the sub-scan unit (including the justifi-cation unit).
㜀㔱

send me an email address I can send you the manual

Log inor Register

to rate and post a comment

(7)

Dario Briceno

re: Error code E 4512 or E4511

July 18, 2017 07:46

Film remains at the V-SYNC sensor. 11/12 depends on which film size is processing. V-SYNC sensor is optical sensor triggered by a phisical flag, so, check the proper function of sensor aand its flag when film is passing by. Check conections and function of transportation motor and Mecanical control board. V-SYNC is located at the entrance of the HPRO unit (just before the laser unit). Open right cover , where you find the film when the errors ocurrs and you`ll find it at the inside top.

Log inor Register

to rate and post a comment

(4)
(1)

ObiwanKenobi

re: Error code E 4512 or E4511

July 18, 2017 07:46

Film remains at the V-SYNC sensor. 11/12 depends on which film size is processing. V-SYNC sensor is optical sensor triggered by a phisical flag, so, check the proper function of sensor aand its flag when film is passing by. Check conections and function of transportation motor and Mecanical control board. V-SYNC is located at the entrance of the HPRO unit (just before the laser unit). Open right cover , where you find the film when the errors ocurrs and you`ll find it at the inside top.

Log inor Register

to rate and post a comment

(7)

Dario Briceno

re: Error code E 4512 or E4511

July 18, 2017 07:46

Film remains at the V-SYNC sensor. 11/12 depends on which film size is processing. V-SYNC sensor is optical sensor triggered by a phisical flag, so, check the proper function of sensor aand its flag when film is passing by. Check conections and function of transportation motor and Mecanical control board. V-SYNC is located at the entrance of the HPRO unit (just before the laser unit). Open right cover , where you find the film when the errors ocurrs and you`ll find it at the inside top.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | US, UK & Europe. Diagnostic Imaging Sales & Service

We supply, service, and support ultrasound, MRI, CT, X-ray, C-arm, and more: new, refurbished, or rental. Serving NHS Trusts, private hospitals, imaging centres across the US, UK, and Europe. ISO 13485 certified. Call: 0333 004 7657 | probomedical.co.uk

stats
advertisement

Upcoming Auction Events

DeoCorp
TIMED
Jul 30, 2026 (229 lots)
ZRG Medical
TIMED
Jul 31, 2026 (229 lots)
Cbrx2sales
TIMED
Aug 04, 2026 (272 lots)