The error codes are explained while using the service software that is provided with every Siremobil Compact. If you need a specific error code explained, and you can't use your service software for some reason, email me & I will look up the error code for you. We also have Siremobil Compact parts in stock, if you need them.
--------------------------
Wayne M. Horsman, VP
Columbia Imaging, Inc.
9050 Red Branch Rd., Ste. O
Columbia, MD, USA 21045
(410) 730-4023
www.columbiaimaging.com
wayne@columbiaimaging.com
4028 is usually a bad Memoskop, but occasionally can be a bad connection on the SG cable connection at X-10 (the main interface cable connecting to the c-ram)
Error E4028 Memoskop unexpected acq end signal
Possible trouble: Host registered unexpected acq end signal of the Memoskop.
Possible help: Check faults on the Memoskop acquisition cable (D1.X10.27).
Check Memoskop.
Check connection cables to the Memoskop.
5006 is an indication that S2 is set to "off" on the D-1 PC Board. (this can happen accidentally during troubleshooting, or during R/R of the d-1 board, etc.)
Otherwise, it may be a "glitch", if it does not recur.
Error E5006 Gen. tried X-Ray on without SS relais
Possible trouble: Generator attempts radiation with SS relay switched off.
Possible help: D1, switch S2 in wrong position.
D1 defective.
Hi Wayne, Thanks so much for taking the time to assist. We may have the issue resolved. Your suggestions led us in the correct direction.
Sincerely,
Joe
How to use the service software of the Siremobil Compact L. We encounter errors while booting the machine, E4021 and E4024. Some errors appears some times E4081. What does it means.
Hello Mr. Wayne,
would you be so kind to send me also the error code list for siremobil compact? My email - markoshv@ukr.net
Thanks in advance for your time for me!
One call to 440-263-6575 and your move will be covered.
We repair & store trailers. We build to your custom needs.
JDI Mobile Medical Manufacturing. Inc. We have your back.
error code siemens siremobil compact
November 07, 2013 03:14
hi Dear all i need error code of siemens siremobil compact c-arm with circuit diagram any one help me msohail4007@yahoo.com
to rate and post a comment
Wayne Horsman
re: error code siemens siremobil compact
November 07, 2013 06:44
The error codes are explained while using the service software that is provided with every Siremobil Compact. If you need a specific error code explained, and you can't use your service software for some reason, email me & I will look up the error code for you. We also have Siremobil Compact parts in stock, if you need them.
--------------------------
Wayne M. Horsman, VP
Columbia Imaging, Inc.
9050 Red Branch Rd., Ste. O
Columbia, MD, USA 21045
(410) 730-4023
www.columbiaimaging.com
wayne@columbiaimaging.com
to rate and post a comment
m sohail
re: re: error code siemens siremobil compact
November 08, 2013 01:16
the error is 4710 4711 4712 4713 4714 4715 4716 4717 4718 and 9102
to rate and post a comment
Peter Russell
re: re: re: error code siemens siremobil compact
November 08, 2013 06:08
These are all errors on the memoscop
to rate and post a comment
Wayne Horsman
re: re: re: error code siemens siremobil compact
November 09, 2013 07:44
Bad Memoskop & slot diaphragm (p/o collimator).
to rate and post a comment
Joe DeJohn
re: re: error code siemens siremobil compact
May 19, 2015 01:48
hello, would you be able to check on the following error codes: 4026 and 5008 Regards, Joe DeJohn
to rate and post a comment
Wayne Horsman
re: re: re: error code siemens siremobil compact
May 19, 2015 10:03
Joe -
4028 is usually a bad Memoskop, but occasionally can be a bad connection on the SG cable connection at X-10 (the main interface cable connecting to the c-ram)
Error E4028 Memoskop unexpected acq end signal
Possible trouble: Host registered unexpected acq end signal of the Memoskop.
Possible help: Check faults on the Memoskop acquisition cable (D1.X10.27).
Check Memoskop.
Check connection cables to the Memoskop.
5006 is an indication that S2 is set to "off" on the D-1 PC Board. (this can happen accidentally during troubleshooting, or during R/R of the d-1 board, etc.)
Otherwise, it may be a "glitch", if it does not recur.
Error E5006 Gen. tried X-Ray on without SS relais
Possible trouble: Generator attempts radiation with SS relay switched off.
Possible help: D1, switch S2 in wrong position.
D1 defective.
- Wayne H.
to rate and post a comment
Joe DeJohn
re: re: re: re: error code siemens siremobil compact
May 20, 2015 10:01
Hi Wayne, Thanks so much for taking the time to assist. We may have the issue resolved. Your suggestions led us in the correct direction.
Sincerely,
Joe
to rate and post a comment
Salvador III Gaviola
re: re: error code siemens siremobil compact
February 05, 2019 08:42
Hi Mr. Wayne,
How to use the service software of the Siremobil Compact L. We encounter errors while booting the machine, E4021 and E4024. Some errors appears some times E4081. What does it means.
Thank ypu
to rate and post a comment
Wayne Horsman
re: re: re: error code siemens siremobil compact
February 06, 2019 08:47
what is your email? I will send you the error code list. - Wayne
to rate and post a comment
Marko Shevchuk
re: re: re: re: error code siemens siremobil compact
February 10, 2019 01:33
Hello Mr. Wayne,
would you be so kind to send me also the error code list for siremobil compact? My email - markoshv@ukr.net
Thanks in advance for your time for me!
to rate and post a comment
Jorge Arevalo
re: re: re: re: error code siemens siremobil compact
August 25, 2022 10:25
Maybe can you help with the service software,? I have the versión VB01F but I can connect with the equipment
to rate and post a comment