I have Picker PQ 2000 CT Scanner.The machine is working and warming up regularly.When I enter the ERP,following caution shows on the screen.
Error Code:1404
Because of this fault,we dont take any images.When I reloaded software,same problem continues.I checked all documents and I didnt find any solution.
Can you please give information regarding this problem
You are not making proper sense with your question. First you say that it is working properly. Then you say you have an error code 1404 (* which does not exist). Are you sure you are not getting an error 1004, which would not allow you to scan? A 1004 is a "small drives error", which is either the beam filter / compensator/ beam shutter / beam limiter, drives not positioning properly, or the related position sensor is bad.
- Dave
Dave,
The machine is working properly.But when I get pilot image,the image is very bad and it has artifacts.
When I enter ERP,following caution shows on the screen
"ECODE 1404 TRACER 1003
VOL8_FILE V8_ALLOCATED non-exit"
If you give me your mail adress,I can send you picture of images regarding this problem.
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Ercan TUTUNCUOGLU
Error Code:1404 Picker PQ 2000 CT Scanner
July 14, 2009 11:23
I have Picker PQ 2000 CT Scanner.The machine is working and warming up regularly.When I enter the ERP,following caution shows on the screen.
Error Code:1404
Because of this fault,we dont take any images.When I reloaded software,same problem continues.I checked all documents and I didnt find any solution.
Can you please give information regarding this problem
Best Regards,
Ercan Tutuncuoglu
to rate and post a comment
Dave Adams
Re: Error Code:1404 Picker PQ 2000 CT Scanner
July 14, 2009 02:23
You are not making proper sense with your question. First you say that it is working properly. Then you say you have an error code 1404 (* which does not exist). Are you sure you are not getting an error 1004, which would not allow you to scan? A 1004 is a "small drives error", which is either the beam filter / compensator/ beam shutter / beam limiter, drives not positioning properly, or the related position sensor is bad.
- Dave
to rate and post a comment
Ercan TUTUNCUOGLU
Re: Error Code:1404 Picker PQ 2000 CT Scanner
July 15, 2009 07:50
Dave,
The machine is working properly.But when I get pilot image,the image is very bad and it has artifacts.
When I enter ERP,following caution shows on the screen
"ECODE 1404 TRACER 1003
VOL8_FILE V8_ALLOCATED non-exit"
If you give me your mail adress,I can send you picture of images regarding this problem.
Best regards,
Ercan
to rate and post a comment
Sergio Leonardo Barral
Re: Error Code:1404 Picker PQ 2000 CT Scanner
July 16, 2009 11:20
such error is related to the spiral disk, not related to your artifact problem. Send me the images please
wickham@fibertel.com.ar
best regards
Sergio
to rate and post a comment
Alexander Tonev
Re: Error Code:1404 Picker PQ 2000 CT Scanner
July 15, 2009 05:02
You reloaded the software with pity or you made a fresh install?
_dita is working or not?
to rate and post a comment
Ercan TUTUNCUOGLU
Re: Error Code:1404 Picker PQ 2000 CT Scanner
July 16, 2009 08:18
Dita doesn't work.
Ercan
to rate and post a comment