Yojin Ramones
error cr35 20C8A
March 19, 2016 12:54
I have a CR 35 in the display appear imaging artifact like a jamme printer paper stuck, when it have not constant moviment. the artefact is vertical.
error appear 20C8A in auto test
to rate and post a comment
juan pablo amado
re: error cr35 20C8A
March 21, 2016 09:25
20C8A
E_SCN_IHW_SWP2
Meaning : Starting the polygon and setting the laser timing failed.
Reason : Probably bad hardware.
Cure : Check previous errors with REM tool.