Posted by an Anonymous User

error CT DX/i

August 18, 2016 12:37

any one help me with this error on GE CT/i:
1.axial scan error code :20-1001-23 error detected by :OGP HV ON was not activated byJEDI on axial scan after EXPCMD was activated byOC
2.XG error phase is 4 code is 60.detail error code:60-0313H exposure error:ILR max current detected

Log inor Register

to rate and post a comment

(12)
(1)

steven

re: error CT DX/i

August 19, 2016 08:09

You must run no lode test to check HV-tank for IRL max problem

Log inor Register

to rate and post a comment

(102)
(13)

Joel Grullon

re: error CT DX/i

August 19, 2016 08:11

The first error is normally generated by another , in your case is clear that is the jedy generator . Try doing service - diagnosis - offline scan and set lower techniques and stationary scan and check if the problem disapear your cause could be the ht transformer arcing

Log inor Register

to rate and post a comment

(7)
(2)

Jogi

re: error CT DX/i

August 19, 2016 08:11

send error log pic to me at addoneinfotech@gmail.com i can help

Log inor Register

to rate and post a comment

(39)
(8)

Raul Elizalde

re: error CT DX/i

August 22, 2016 10:08

You have to make HV tank Diagnostics without load, the problem may be is the tank or bubbles in the tube, if you don't' know how to make that send me email and explain you.

Log inor Register

to rate and post a comment