hallo Dears
i getting an error in CT SCAN GE Prospeed
(X.Ray Zero detect LCP-03 TN-SEQ ADD-0024 EEEC)
can you told what's this error and who to solve this problem
many thanks
Dear Ahmed
i don't have expereice in this CT but i am CT eng. since 1996 i hope i can help you.
when this error appears during x ray, during prepration ...etc?
ahmed buaziza
Error CT SCAN GE Prospeed
July 21, 2010 10:16
hallo Dears
i getting an error in CT SCAN GE Prospeed
(X.Ray Zero detect LCP-03 TN-SEQ ADD-0024 EEEC)
can you told what's this error and who to solve this problem
many thanks
to rate and post a comment
Borhan Kalash
Re: Error CT SCAN GE Prospeed
July 23, 2010 10:04
Dear Ahmed
i don't have expereice in this CT but i am CT eng. since 1996 i hope i can help you.
when this error appears during x ray, during prepration ...etc?
to rate and post a comment
Carlos Duran
Re: Error CT SCAN GE Prospeed
August 20, 2010 08:48
this appears in two scenarios:
system ok: because the patient is too fat, and acquisition was made using too low radiation
solution: raise exposure.
system bad: there is no data from gantry, no communication, or DAS problem
you must check data path.
to rate and post a comment
Claude Coelo
Re: Error CT SCAN GE Prospeed
December 27, 2010 01:44
Maybe i can help you
Claude
to rate and post a comment