sunil Kumar
Error E190 GE 600T
October 24, 2005 02:56
Any details on GE600T mammo error 190.manual does not show this error.
to rate and post a comment
Manuel Casanova
Re: Error E190 GE 600T
October 24, 2005 08:51
Mr Kumar:
Follow this steps:
1.- There where transiets on the line supply, inintialization took place, check this: push OFF and then ON again.
2.- Switch OFF senographe and place the 400PL7 board on extender, check: optocoupler B2, the voltage measured be 0.09V +- 50 mv, Led D4 should be lit and +5v of Lambda powewr supply should be present on 400PL7, if required replace optocouple B2.
God luck