Error, Elscint select sp

(63)
(11)

Ronnie Aguiar

Error, Elscint select sp

September 29, 2009 09:35

error: "heat exchange is off when must in on" i need help please

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

Re: Error, Elscint select sp

September 29, 2009 04:30

When this problem appears? During scanning?
Heat exchanger is powered from AMC unit but monitored by SCU. Sometimes, when HV spikes there are some strange errors.
If it is not during scanning check if Heat Exchenger is powered.
Open the front cover, check if the fans in HExchanger are working (air is flowing). If not - press reset button on SCU. If fans are still not running and check if leds on SCU are lit. If not check the fuses on AMC module (switch off the power to rotor first! - on PDU - right leg).

I am afraid it is during scanning. In this case it is not easy to say what exactly is failed.
Heat Exchaner on/off signal is flowing from this module through SCU. HV spikes can disturb this signal. Check carefully all 4 HV connections and reconnect them using silicone oil. There should be no air in this connections. It sometimes helps. If you need any details send me a message.

Log inor Register

to rate and post a comment

(63)
(11)

Ronnie Aguiar

Re: Error, Elscint select sp

September 30, 2009 09:34

heat exchange is off when must on,This error appears when shooting with the following parameters 120 kv 60 ma time 2 seconds, is there any test that can be performed to detect this problem, many thanks for help



please help -me

ronnie aguiar

brazil

Log inor Register

to rate and post a comment

(63)
(11)

Ronnie Aguiar

Re: Error, Elscint select sp

September 30, 2009 09:30

heat exchange is off when must on,This error appears when shooting with the following parameters 120 kv 60 ma time 2 seconds, is there any test that can be performed to detect this problem, many thanks for help



please help -me

ronnie aguiar

brazil

Log inor Register

to rate and post a comment

(2)

(2)

Pablo Aguayo

re: Re: Error, Elscint select sp

November 02, 2016 10:34

Mr. Selwesiuk:

I hava the error: ER80, 3B15,2,78,19,,19,18,1,13F8,7980.
CT Scanner Elscint Select SP-
Can you help me?

What do you think can be the problem?

Thank you very much.

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

re: re: Re: Error, Elscint select sp

November 03, 2016 06:00

It is HV tolerance error. Please describe more details: when it happens, what are scan parameters. I recommend to use SCU_diag login and continue. Remember to change diagnostics switch position on RMS&COMM. For photos please write email.

Log inor Register

to rate and post a comment

(44)
(9)
(1)

Roberto Villalba

re: re: re: Re: Error, Elscint select sp

November 11, 2016 10:07

Hello:

How I can be sure if the X-ray tube is broken?

I have error:

error from SCU 3CA8,
or error from SCU 3B15

Thank you

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

re: re: re: re: Re: Error, Elscint select sp

November 13, 2016 12:18

3CA8 means - Emission current tolerance,
3B15 - HV tollerance. The best way is to check HV behaviour in SCU_diag login. Then you can observe all readings (voltages, currents). Do you know how to do it? BTW, is your system going to be in ready state before scan? You can write on email also for deatiled discussion

Log inor Register

to rate and post a comment

(44)
(9)
(1)

Roberto Villalba

re: re: re: re: re: Re: Error, Elscint select sp

November 14, 2016 12:59

The system reach the ready state for scan, but when press the botton to make the scan, the error appears. Thank you - (always I can see the yellow ligth of the IPC module (fault) but the 540 V in the rinds are presents.

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

re: re: re: re: re: re: Re: Error, Elscint select sp

November 17, 2016 01:09

Your explanation means that your IPC doesn't go to normal condition. What is happening when you reset IPC module? Can you see fault led on all the time? Check 50A fuses - maybe one phase is missing? Do you have an access to scu_diag login?

Log inor Register

to rate and post a comment

(44)
(9)
(1)

Roberto Villalba

re: re: re: re: re: re: re: Re: Error, Elscint select sp

November 17, 2016 02:46

Always the faul light in the IPC is ON. But the 540V DC is OK.

Log inor Register

to rate and post a comment