Daniel de Jesus Aguirre Torres

Error GE HiSpeed

February 15, 2019 09:10

Some ona can helpme i need to know whta is this errors pleas.

*Cradle Error. Code: 12-1003-03 Error detected by : TGP (TP) Cradle slip was detected in response to CRADLE CMD.

* XG Error phase is 2 cde is 90. Detail error code : 90 - 0701H Application error : Saved RAM checksum pb

* XG Error phase is 1 cde is 90. Detail error code : 90 - 0701H Application error : Saved RAM checksum pb

Log inor Register

to rate and post a comment

(32)
(2)

cesar lizarazo

re: Error GE HiSpeed

February 16, 2019 06:19

90–070
1H
NV RAM checksum pb.
Generator kV control
board has detected cor-
ruption in the NV Ram
verification (After power
up)
1. Database problem
2. External cause
(Spikes)
3. kV control
If persistent :
–Reload the Data base. If no im-
provement :
–Replace kV control board and re-
load the data base

Log inor Register

to rate and post a comment

(102)
(15)
(1)

Maurice Dilick

re: Error GE HiSpeed

February 17, 2019 11:10

The data in your NV RAM on the KV Control Board is incorrect, if you have a stored system, upload and, unless the KV Control Board has failed, should be fine.

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: Error GE HiSpeed

February 17, 2019 11:19

Please see below:

12-1003-03 error is caused by patient movement during cradle movement or the cradle encoder is faulty.
If error did not occur during patient scanning, then check cradle encoder or cradle roller for any foreign objects.

90-0701H is a database problem were the KV control detects a database corruption on the NVRAM.
To resolve this, please first re-install the database using the JEDI service tool kit CD.
If this fails, replace the KV Control board and re-install database.

craigwambura@outlook.com
+256 782 702243 (WhatsApp)

Log inor Register

to rate and post a comment

stats
DOTmed text ad

General Biomedical Service - New & Refurbished Ventilators

GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449

stats

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)