Error in Acuson Sequoia C512

Tags: Ultrasound, ACUSON SIEMENS

Blend Fathullah

Error in Acuson Sequoia C512

March 05, 2014 04:21

Dear All
I have a problem in my ultrasound Sequoia C512 when I switch on the system boot normally in the first and after that the boot stopped and give me this error 0-0-2
please any one can help me to solve this problem

Thanks

Salus Medical Trading Company
Salusmedico@gmail.com

Log inor Register

to rate and post a comment

(4)

Marty Bayless

re: Error in Acuson Sequoia C512

March 05, 2014 06:51

You have a bad RDP board. Replace and you should be fine.

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: Error in Acuson Sequoia C512

March 05, 2014 06:51

Do you have the service manual?

Log inor Register

to rate and post a comment

(93)
(54)

mohamed elgenidy

re: re: Error in Acuson Sequoia C512

March 05, 2014 08:38

I have it

Log inor Register

to rate and post a comment

(6)
(1)

Jeff Majors

re: Error in Acuson Sequoia C512

March 06, 2014 01:34

If you pulled the side covers off the card cage try rebooting a few times and error should go away if you are at 9.51 or 12.221 sw rev. If fans run high this is another indication of this issue.

Otherwise you might have rdp or board that has failed. If you have
Hyperterminal try booting again and respond with boot sequence and any put failures.

Jeff

Pinnacle medical imaging

Log inor Register

to rate and post a comment

(3)

Alfred Kuqali

re: re: Error in Acuson Sequoia C512

May 10, 2016 04:49

Occasionally, in my Acuson Sequoia C512 happened an interruption of the work, picture in display and all the keys of keyboard were out of the work. Turning off could be done only by pulling of the plug from the socket.
HDD resulted as corrupted. The old HDD was replaced with a new HDD with the same parameters to the existing one. Installing SW and back-up were done.

Now, during the normal work of ultrasound machine, occasionally again happen interruption of work, all the keys of keyboard are out of work. The clock of machine is stopped. This failure happens during procedure of starting or in the end of this or during of the normal work or in the pause.
I have never seen workstation behand of monitor working. Maybe this may have any connection with the problem.

Does anybody know what happens with this ultrasound machine?

Log inor Register

to rate and post a comment

(100)
(9)
(4)

JORGE LUIS LOPEZ

re: re: re: Error in Acuson Sequoia C512

January 29, 2024 10:52

Alfred, i'm working on a similar problem, whit a sequoia c512,
the operator console is non responsive, i cannot turn it off via power button, and it does not show any error on boot up, it shows the image, cannot change probes. it just freezes the console.
did changing the HD. fix ed it?

Log inor Register

to rate and post a comment

(64)
(15)

AHMEDELBAGA

re: Error in Acuson Sequoia C512

March 06, 2014 07:44

Dear Blend
the error massage (0.0.2) and booting to sui not to imaging depend on several things please e-mail me at:
elbaga85@hotmail.com
we will check it setp by step tell we solve it or diagnose the error to solve it.

1 person liked this
Log inor Register

to rate and post a comment

(2)

Diego Armando Serrano Vázquez

re: re: Error in Acuson Sequoia C512

May 19, 2021 10:54

Would you like help me, i have a problem with a Sequoia 512 the monitor don't have imagen, and don't have the service manual only technical and said me the problem is with the boards BDM and RDP

Log inor Register

to rate and post a comment

AHRImedical

re: Error in Acuson Sequoia C512

August 25, 2016 04:36

We are getting the system error that the MO Disk was not ready after just booting the system up while it still has the boot up gray screen with picture of the tree and will not go past this system error into the imaging screen, however it was not doing this earlier today and just started doing this. My colleague thinks he hit a button on the underside of the keyboard earlier that reset the system. Since then we have tried to start it up and first got “System Error MO Disk Not Ready.” But now when we boot up the system with the MO Disk in the drive it now shows the system error that “lynx.os.ssp cannot be found.” I was hoping that maybe you guys might have some ideas about what was going on and how we could address this issue to fix it.



Thank you in advance for your help.



Please reply here or email us at andrew@ahrillc.com

Log inor Register

to rate and post a comment

(77)
(18)

Brian Eberling

re: re: Error in Acuson Sequoia C512

September 02, 2016 06:13

Andrew, your hard drive cable either came loose or your hard drive went bad. Open the back panel below the monitor and check all your cable connections. If system still doesn't boot, you will need a new hard drive. Order the exact model drive that is in there and make sure all the jumper settings on the new drive are set exactly the same as the old drive. After reloading software and restoring from backup, you will be back in business.

Log inor Register

to rate and post a comment