Error in Checkup Somatom Esprit

Tags: CT, SIEMENS

(34)
(6)

Kian Chaves

Error in Checkup Somatom Esprit

December 19, 2017 09:16

Hello Guys,
I'm serving a client who has an older Somatom with IRS 5 and ICS 6. The client made a call because the Workstation didn't go up the Syngo. It has been reported that the CT isn't switched on for 7 months. Analyzing the IRS and ICS I found that they were running out of battery and losing the date / time. After changing the battery and setting the date and time the system has turned on perfectly. However at the time of checkup I got the following message:
Initilization failed.
Filament adaptation aborted by error.
And by viewing the Event Log the following message was reported
Invalid / inconsistent value: Cooling duration is negative! Probaly the systemtime has been changed or tubehistory is corrupt.
May someone help me?

1 person liked this
Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: Error in Checkup Somatom Esprit

December 19, 2017 03:56

Hm i think you need find tube history file on C disc. And delite all from this folder. Then reboot system

1 person liked this
Log inor Register

to rate and post a comment

(34)
(6)

Kian Chaves

re: re: Error in Checkup Somatom Esprit

March 06, 2018 02:17


I'll look for the tubehistory file and try this possible solution. Thank you very much for the information I will be starting the service now

Log inor Register

to rate and post a comment

(94)
(22)

Mario E Sarduy La Paz

re: Error in Checkup Somatom Esprit

December 20, 2017 09:15

Hi.
Maybe doing a restore the CT setting could be other solution too.
Best Regards and let us know the final result or share the eventlog file for do a better analysis of the problem.

1 person liked this
Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: re: Error in Checkup Somatom Esprit

December 20, 2017 03:37

Restore setting restored only system setting, but in this situation system uoad information from tube and thats why he has a error. For test, change date on your sistem which was before replased battery

1 person liked this
Log inor Register

to rate and post a comment

(34)
(6)

Kian Chaves

re: re: re: Error in Checkup Somatom Esprit

March 06, 2018 02:17


I'm gonna try now. I'll give you a return as soon as possible, thank you very much for the help

Log inor Register

to rate and post a comment

(34)
(6)

Kian Chaves

re: re: re: Error in Checkup Somatom Esprit

March 09, 2018 03:53

The IRS motherboard had a problem with a capacitor near the battery and CMOS, I made the switch and it's OK. As for the tubefile file I have already found it however I don't have access to delete it. What should I do to gain access?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

JOHNSON & JOHNSON MONARCH QUEST - For Sale $299k or Lease $6k Per Month, 48 Months

Lightly used system available direct from the end user customer. MONARCH QUEST Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)