Punit Singh

error in offset data acquisition,asteion multi

May 26, 2015 07:20

I am getting error in offset data acquisition.error 13 bit is 45 rest are 40.have changed all rot PCB but situation remain same .earlier it was intermittent but now permanent

Log inor Register

to rate and post a comment

(38)
(7)
(1)

Levan Gamtsemlidze

re: error in offset data acquisition,asteion multi

May 28, 2015 06:30

Hi,

I have the same problem with two different asteion multi scanners, in one case problem was in EWS300-5 5V @ 60A Power Supply 410W unit which is located under the detector, in second case was damaged GCIFM board which located right side of Gentry stationary part, near KGTSM board.

good luck