Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Error in XC.

Tags: CT, TOSHIBA

(5)
(1)
(1)

Mesayon Kaewsena

Error in XC.

February 18, 2016 08:53

I have error from CT Alexion 16 Slice show below.

Error in XC.
code [0xb2640000] dev[0x00010002] Pno[0]
---------------------------------------------------------

---ERROR INF dump---
SSADI STATUS
ss_state [0x0010], valid_bit[0x000f]
GTS STATUS
mode[OR], exec[ED], error[0x00000000]
InterLock Slide[00], height[00], tilt[00]
InterLock LRside[00]
free&light[00], out2censor censor[00]
MUDAT status[0x0000]
censor information
3032303030303030
GTS DETAIL STATUS
detail error
40404040404040404040404040404040
40404040404040404040404040404040
DDmotor detail | error
30303030303030303030303030303030
slide error detail
3030303030000000
height error detail
303030303030303030303000
L/R side error detail
000000000000000000000000
l mudat error information
30303030303030303030303030303030
30303030303030303030303030303030
30303030303030303030303030303030
XC STATUS
mode[OR], exec[DR], error [0x1007f7d0]
XC DETAIL error
3f3f3f3f3f3f3f3f3f3f2f2f2f2f2f00

Thank for your advise
Toon
mka@informedgroup.net



1 person liked this
Log inor Register

to rate and post a comment

(18)
(5)
(1)

Jorge Barrios

re: Error in XC.

February 19, 2016 08:46

Check 24 v power supply, it is under HV unit.
If this is ok, you will have to change XC board.
Good luck!!!

1 person liked this
Log inor Register

to rate and post a comment

(5)
(1)
(1)

Mesayon Kaewsena

re: re: Error in XC.

February 19, 2016 09:02

Thank you for your advise

Log inor Register