April 10, 2013 11:15

i am finding an error of 30 and 16 in senographe 600T senix HF. please explain the reason

Log inor Register

to rate and post a comment

(27)
(3)
(5)

James McGinty

re: Error list explanation

April 10, 2013 04:25

Error 30, Ram checksum, probably cause, battery on cpu board. Or restart service program and perform another checksum.

Error 16, do you mean error 160 ? Large focal spot open, 400PL1 or 400 PL2

jim@digitecmedical.com

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

re: Error list explanation

April 12, 2013 09:53

if you need parts or sm please send me email at semmedical@gmail.com