Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Error on CR30-X

Tags: PACS / DR / CR / Digital, AGFA

(6)

Oluwafemi Abati

Error on CR30-X

October 18, 2016 02:28

Error 14870, pls how can I resolve this error.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Error on CR30-X

October 19, 2016 06:43

14870;"PMI_ELABEL";"Error";"";"";"";"Error getting data from PMI eLabel";"Meaning: Error getting data from PMI eLabel.

Reason: PMI eLabel defective.

Cure: Replace PMI board.";"";"";"";

1 person liked this