I need your help to correct ERROR on CT SOMATOM PLUS 4 EXPERT, because when I attemp to WARM-UP, CALIBRATE OR START A STUDY, the equipment STOP and send the following message:
MASME_ABORTED Function aborted
MASME_MTIMER_EXPIRED Master control timer expired (IS, RS or AS)
MAS_E000011 initdone timeout ( 'E' 0 O 11 57 dc f c0 ).
Any advice, is appreciated.
Thanks
Hello julio,
First of all thanks for your comment... . Julio, what do you mean with MCU, are you refering MCU is in one leg of Ct Gantry?; Please give me the fisical location of MCU you are talking about. Could you give me your e-mail; mine is imesamex@prodigy.net.mx
Thanks
Pedro Zamora
ERROR ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 17, 2013 05:46
¡¡HELLO EVERYBODY!!,
I need your help to correct ERROR on CT SOMATOM PLUS 4 EXPERT, because when I attemp to WARM-UP, CALIBRATE OR START A STUDY, the equipment STOP and send the following message:
MASME_ABORTED Function aborted
MASME_MTIMER_EXPIRED Master control timer expired (IS, RS or AS)
MAS_E000011 initdone timeout ( 'E' 0 O 11 57 dc f c0 ).
Any advice, is appreciated.
Thanks
to rate and post a comment
Julio c. Gomez
re: ERROR ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 18, 2013 08:21
Hello Pedro.
may be you have a comm error:
check the stop loop report a mcu is no sync, or you have a card invalid.
host--MCU; MCU--RF transceiver; RF transceive--onboard controller.
good luck.
Julio C. Gomez.
Veracruz, Mexico.
to rate and post a comment
Pedro Zamora
re: re: ERROR ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 18, 2013 05:33
Hello julio,
First of all thanks for your comment... . Julio, what do you mean with MCU, are you refering MCU is in one leg of Ct Gantry?; Please give me the fisical location of MCU you are talking about. Could you give me your e-mail; mine is imesamex@prodigy.net.mx
Thanks
to rate and post a comment
Daniel Gang
re: re: re: ERROR ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 19, 2013 01:06
Yes, mcu monuted on left lside of gantry.
to rate and post a comment