Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

ERROR ON CTSCAN 64 SLICE

Tags: CT, PHILIPS

(4)

Hans Yambazi

ERROR ON CTSCAN 64 SLICE

October 30, 2015 10:12

WHEN DOING AIR CALIBRATION I AM GETTING THE CT "RESOURCE ALLOCATION FAILED" AND WHEN DOING CT PROCEDURE I AM SETTING "GANTRY ACQUISITION ERROR!

I have very limited knowledge on this machine Philips brilliant 64 slices we hardly use it for 3 months and based in Africa -Tanzania please help

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: ERROR ON CTSCAN 64 SLICE

October 31, 2015 11:02

this could be several things, you need to login to the CIRS and check the event log. could be software but would know more after seeing the event log entries

1 person liked this
Log inor Register

to rate and post a comment

(138)
(18)
(1)

umberto sciunnach

re: ERROR ON CTSCAN 64 SLICE

October 31, 2015 11:02

Hello
Are you talking about philips brilliance 64 slices ?
Regards
Umberto

Log inor Register

to rate and post a comment

(4)

Hans Yambazi

re: re: ERROR ON CTSCAN 64 SLICE

November 01, 2015 08:23

yes

Log inor Register

to rate and post a comment

(138)
(18)
(1)

umberto sciunnach

re: ERROR ON CTSCAN 64 SLICE

October 31, 2015 11:03

If yes I can explain the procedure to found the fault.
Regards
Umberto

1 person liked this
Log inor Register

to rate and post a comment

(4)

Hans Yambazi

re: re: ERROR ON CTSCAN 64 SLICE

November 01, 2015 08:16

Dear Umberto
The ct is philips brilliant 64 slice I will appreciate if you send procedure of fixing this problem

Log inor Register

to rate and post a comment

(138)
(18)
(1)

umberto sciunnach

re: re: re: ERROR ON CTSCAN 64 SLICE

November 01, 2015 08:24

Hello
At first check the cpu at the right of ganrty all supply must on.
After that you must begin with computer consolle , try to reconstruct some images.
Let me know .
My mail umbertosc1957@gmail.com
Regards
Umberto

Log inor Register

to rate and post a comment

(105)
(17)
(17)

Doscar Ernesto Perez Pereyra

re: ERROR ON CTSCAN 64 SLICE

October 31, 2015 11:03

Resource Allocation is a typical problem when the gantry not know the info received by the host.


But, GANTRY ACQUISITION ERROR is at problem in acquision system (RCOM, ACQ, CIRS).

You need enter in the CIRS Server 1 (Where is located the Acquisitor 4.5 GB) and run test BUT BEFORE DO TEST you need closed the rj45 loop with LAN2 and put the FIBER OPTIC LOOP. (IF YOU NOT PUT THE LOOPS THIS TEST WILL FAIL).


Later enter to SERVER 2 an run test in 3DBP for be sure the Backprojector are good. (This not need any loop).


When you do that please talk me by email: doscarperez@gmail.com

Log inor Register

to rate and post a comment

(75)
(8)
(7)

JIFFY PAUL AKKARAKARAN

re: ERROR ON CTSCAN 64 SLICE

October 31, 2015 11:05

Please send the model no and the complete details of the CT

Log inor Register

to rate and post a comment

(4)

Hans Yambazi

re: re: ERROR ON CTSCAN 64 SLICE

November 06, 2015 02:29

when I open event viewer I got error code no 33227156 on gantry please help

Log inor Register

to rate and post a comment

Dustin Zimmerman

re: ERROR ON CTSCAN 64 SLICE

October 31, 2015 11:06

Could be a number of things. If you run Save the logs from application and email them to me I can try to help. Those are generic errors that operator sees. Need to see the actual error logs. Email dzimmerman@rsti-training.com

Log inor Register