Manuel Casanova

ERROR ON LUNAR NT

June 27, 2012 01:09

HI: I HAVE A ERRRO ON LONGITUDINAL STEPS ARE 1.353,25 ALL OTHER TEST IS WORKING OK, BUT LONGITUDINAL STEPS IS PROBLEM.

ENCORE 2004, NEED THE PASSWORD

BESTS REGARDS

1 person liked this
Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: ERROR ON LUNAR NT

June 27, 2012 05:27

password could be Award, Smile, badger or madison depending on software version.
Check the drive belt tension, encoders and sensors, if ok check connectors on the FOINK board

2 people liked this
Log inor Register

to rate and post a comment

(37)
(7)
(3)

Albert J. GARRY

re: ERROR ON LUNAR NT

June 27, 2012 11:39

Hello Manuel,

Your problem is the limit switch. Either the switch at the foot end is loose or the head end switch is loose. Check the switch brackets for tightness. To adjust, move either switch to reduce the counts until the counts are within specifications. Look at a previous passing QA to determine where to set the counts. Afterwards, make sure both brackets are tight. Over time, the bracket become loose if they were not properly tightened during the installation. Let me know. If needed, send me your number and I will call you tomorrow. my email....aj.garry@gmail.com

AJ Garry

1 person liked this
Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

re: re: ERROR ON LUNAR NT

June 28, 2012 11:16

THANK“S YOU A LOT

Log inor Register

to rate and post a comment

(37)
(7)
(3)

Albert J. GARRY

re: re: re: ERROR ON LUNAR NT

June 29, 2012 12:10

Hello Manuel,

Was the limit switch the problem? Just let me know so I will know if my data was accurate.

regard,

AJ

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

re: re: re: re: ERROR ON LUNAR NT

July 03, 2012 04:17

ALBERT: the problem is a wire cut from limit foot switch, and check the configuration table and equipment is work 100%, thanks a lot

Log inor Register

to rate and post a comment

(37)
(7)
(3)

Albert J. GARRY

re: re: re: re: re: ERROR ON LUNAR NT

July 03, 2012 07:17

EXCELLENT!!!!! Great job Manuel......

AJ

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

re: re: re: re: re: re: ERROR ON LUNAR NT

July 05, 2012 04:18

Albert: I have a trouble with calibration, error in the spillover 12.25% and counts is low, any ideas??

Log inor Register

to rate and post a comment

(37)
(7)
(3)

Albert J. GARRY

re: re: re: re: re: re: re: ERROR ON LUNAR NT

July 06, 2012 08:41

Hello Manual, is this the same system that was working 100% after you located the broken wire with the foot end limit switch? If, so, by calibration, do you mean the daily QA? If so, the QA block is not set up correctly...either it is upside down or facing the wrong way. When is a good time for us to talk so I can get a better feel for your system? Regards, AJ

Log inor Register

to rate and post a comment