ERROR ON MX-8000

(33)
(7)
(11)

Olga De La Paz

ERROR ON MX-8000

November 18, 2011 03:36

CAN SOMEONE HELP ME WITH THIS ERROR ON A PHILIPS MX 8000
THE UNIT DOES THE WARM UP AND ALL BUT WHEN ITS GOING TO INITIATE WE GET GENERAL ERROR MESSAGE
AND THEN 34209825 S_MASTER_SEQUENCER_ACQ_NOT_OK

WE DON’T HAVE THE MANUALS .

Log inor Register

to rate and post a comment

(24)
(1)

sarmad hana

Re: ERROR ON MX-8000

November 18, 2011 04:43

check fiber optic cable from gantry to computer

Log inor Register

to rate and post a comment

(62)
(11)
(14)

James Fower

Re: ERROR ON MX-8000

November 18, 2011 08:36

I worked once on these systems, the fiber optics is a good idea.
As well as the fiber optics connections on the gantry side, these systems have 2 FO receivers and multi optics transmitters in the rotating gantry, in such that at least one is on at time. Good luck.

Log inor Register

to rate and post a comment

(20)

BINH PHAN

Re: ERROR ON MX-8000

November 18, 2011 10:21

Not data aquision!
check sign from DMC to Recon or Control link.

Log inor Register

to rate and post a comment