Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Error pc01 psc . mri 5000ad hitachi

Tags: MRI

abdul ghani shah

Error pc01 psc . mri 5000ad hitachi

May 22, 2013 07:53

Please guide me

Log inor Register

to rate and post a comment

(3)

Aleksandar Kanacki

re: Error pc01 psc . mri 5000ad hitachi

May 23, 2013 08:22

PC 01 PSC timeover detected. Power ON again. Retry.

This is the error in PSC (pulse sequence controller) PCB in the console. Run PSC check memory test on the console (from Check mode). From the results you should know is the PSC board defective itself or it's some memory board.