Over 150 Missouri Auctions End Today - Bid Now

Error polydoros lx 30/50

Tags: Radiology - R/F

(53)
(6)

German Mamani

Error polydoros lx 30/50

December 04, 2018 09:56

Some body knows about error 008/240
Before o after the system show other code error like 007/034 or 007/051
Some suggestions about it
Thanks

Log inor Register

to rate and post a comment

(109)
(15)
(1)

Svilen Popov

re: Error polydoros lx 30/50

December 04, 2018 05:57

Error 007
Meaning:
Polydoros is no longer in the XCS-network.
Error 007 is most of the times accompanied by a 1 on the D100 in case of HW defects.
In case of SW-Resets 007 will heal itself.
Description:
The communication between Polydoros and XCU is down.
Possible reason:
• Checking the terminal connectors. At the terminal D320.X3.S and US or D320.X4.S
and US, there must be either:
- 2 terminal connectors installed
- 1 terminal connector and 1 cable installed
- 2 cables installed
The terminal connector has Part No.: 3157174.
• Cable not connected from XCU to generator
• Cable has a short circuit
• Hybrid-circuit of ARCNET is defective on any XCS component
• SW of a component sends ARCNET Reconfiguration interrupts intermittently
• Watchdog of Polydoros is active
• D100 or power supply defective
Possible causes and action:
• Check cabling for connection and shorts
• Check Hybrids and terminators.
• D100, D200 defect, replace.

Check cables from power supply to D100 and D200, and reseat them

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)