Error 007
Meaning:
Polydoros is no longer in the XCS-network.
Error 007 is most of the times accompanied by a 1 on the D100 in case of HW defects.
In case of SW-Resets 007 will heal itself.
Description:
The communication between Polydoros and XCU is down.
Possible reason:
⢠Checking the terminal connectors. At the terminal D320.X3.S and US or D320.X4.S
and US, there must be either:
- 2 terminal connectors installed
- 1 terminal connector and 1 cable installed
- 2 cables installed
The terminal connector has Part No.: 3157174.
⢠Cable not connected from XCU to generator
⢠Cable has a short circuit
⢠Hybrid-circuit of ARCNET is defective on any XCS component
⢠SW of a component sends ARCNET Reconfiguration interrupts intermittently
⢠Watchdog of Polydoros is active
⢠D100 or power supply defective
Possible causes and action:
⢠Check cabling for connection and shorts
⢠Check Hybrids and terminators.
⢠D100, D200 defect, replace.
Check cables from power supply to D100 and D200, and reseat them
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
German Mamani
Error polydoros lx 30/50
December 04, 2018 09:56
Some body knows about error 008/240
Before o after the system show other code error like 007/034 or 007/051
Some suggestions about it
Thanks
to rate and post a comment
Svilen Popov
re: Error polydoros lx 30/50
December 04, 2018 05:57
Error 007
Meaning:
Polydoros is no longer in the XCS-network.
Error 007 is most of the times accompanied by a 1 on the D100 in case of HW defects.
In case of SW-Resets 007 will heal itself.
Description:
The communication between Polydoros and XCU is down.
Possible reason:
⢠Checking the terminal connectors. At the terminal D320.X3.S and US or D320.X4.S
and US, there must be either:
- 2 terminal connectors installed
- 1 terminal connector and 1 cable installed
- 2 cables installed
The terminal connector has Part No.: 3157174.
⢠Cable not connected from XCU to generator
⢠Cable has a short circuit
⢠Hybrid-circuit of ARCNET is defective on any XCS component
⢠SW of a component sends ARCNET Reconfiguration interrupts intermittently
⢠Watchdog of Polydoros is active
⢠D100 or power supply defective
Possible causes and action:
⢠Check cabling for connection and shorts
⢠Check Hybrids and terminators.
⢠D100, D200 defect, replace.
Check cables from power supply to D100 and D200, and reseat them