Errors on CT Max 640

July 03, 2007 08:00

During scanning on CT Max 640 there had been switching off of the power.After switching on of CT the following errors appeared:
*AL-108-Table Limit
*ER-121- Gantry fail 0004
*ER-101- TOP Time out

Who knows what do above errors mean?
Best regards,
MEDIVIT

Log inor Register

to rate and post a comment

(82)
(19)

Uldis Verdins

Re: Errors on CT Max 640

July 03, 2007 12:52

Gantry rotor will be in limit position.
You need to remove rear side cover, turn back gantry (using motor spindle) and reactivate limit swich. Power on system and look what happen. Good lock.