During scanning on CT Max 640 there had been switching off of the power.After switching on of CT the following errors appeared:
*AL-108-Table Limit
*ER-121- Gantry fail 0004
*ER-101- TOP Time out
Who knows what do above errors mean?
Best regards,
MEDIVIT
Gantry rotor will be in limit position.
You need to remove rear side cover, turn back gantry (using motor spindle) and reactivate limit swich. Power on system and look what happen. Good lock.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Victor Grebenuyk
Errors on CT Max 640
July 03, 2007 08:00
During scanning on CT Max 640 there had been switching off of the power.After switching on of CT the following errors appeared:
*AL-108-Table Limit
*ER-121- Gantry fail 0004
*ER-101- TOP Time out
Who knows what do above errors mean?
Best regards,
MEDIVIT
to rate and post a comment
Uldis Verdins
Re: Errors on CT Max 640
July 03, 2007 12:52
Gantry rotor will be in limit position.
You need to remove rear side cover, turn back gantry (using motor spindle) and reactivate limit swich. Power on system and look what happen. Good lock.
to rate and post a comment