Hi everybody, I need your help to correct CT SCANNER SOMATOM PLUS 4 EXPERT OF SIEMENS, when WARM-UP, CALIBRATION or SHOOT process attempt to carry-on.
The errors are the following:
1.- MASME_ABORTED Function aborted
2.- MASME_MTIMER_EXPIRED Master control
3.- MAS_E000011 initdone timeout ('E' 0 0 11 57 dc f c0)
I will appreciate any advice about.
Thanks in advance
Does the system start to expose and stop or dont stop at all?
Try something easy first.
Reload the f/ware to the controllers.
Check the error log for any information and help and speak again
Check carefully the errors. before this sould be apear more with info inside.
Check if after the exposure fail the system comes to standby.give it litle time for this
re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 19, 2013 04:36
George,
Regarding you are asking me, the system NEVER is ready to start , because the GANTRY has ONLY the small lights of its operation front panel and the error log show error that initially is in the POST.
Could you please send me the procedure for RELOADING the f/ware to the controllers?
Thanks.
re: re: re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 24, 2013 07:27
Dear George,
Let me tell you that, at the moment to check small display, it does not show anything to help me for finding any info of board reported. But actually, when system start with WARM-UP operation send following code sync_lost_ rs and stop rotation of gantry.
Waiting your comments
Thanks
re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
April 17, 2014 01:15
Hi this is Valentin Cardona
My Plus 4 expert give error MAS_700013 when init the warm up
I dopnt have the technical information about errors
Can you help me?
Thanks
Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty!
Email P/N for pricing: sales@Relaxationinc.com
(P) 813-925-1314, www.relaxationinc.com
Pedro Zamora
ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 17, 2013 08:27
Hi everybody, I need your help to correct CT SCANNER SOMATOM PLUS 4 EXPERT OF SIEMENS, when WARM-UP, CALIBRATION or SHOOT process attempt to carry-on.
The errors are the following:
1.- MASME_ABORTED Function aborted
2.- MASME_MTIMER_EXPIRED Master control
3.- MAS_E000011 initdone timeout ('E' 0 0 11 57 dc f c0)
I will appreciate any advice about.
Thanks in advance
to rate and post a comment
Georgie
re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 19, 2013 12:55
Does the system start to expose and stop or dont stop at all?
Try something easy first.
Reload the f/ware to the controllers.
Check the error log for any information and help and speak again
Check carefully the errors. before this sould be apear more with info inside.
Check if after the exposure fail the system comes to standby.give it litle time for this
to rate and post a comment
Pedro Zamora
re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 19, 2013 04:36
George,
Regarding you are asking me, the system NEVER is ready to start , because the GANTRY has ONLY the small lights of its operation front panel and the error log show error that initially is in the POST.
Could you please send me the procedure for RELOADING the f/ware to the controllers?
Thanks.
to rate and post a comment
Pedro Zamora
re: re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 21, 2013 09:00
I dont know how to find which controler is out!!!!
Could you give me your e-mail, this ia mine: imesamex@prodigy.net.com
Thanks
Pedro
to rate and post a comment
Georgie
re: re: re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 21, 2013 09:20
Retype your email.This one is wrong
to rate and post a comment
Vimal Prakash Tripathi
re: re: re: re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
January 18, 2014 08:55
Dear Georgie,
Could you please send me the procedure for RELOADING the f/ware to the controllers in plus 4?
my id is anand_8050@yahoo.com
Thanks.
to rate and post a comment
Georgie
re: re: re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 21, 2013 09:20
Check the email you gave me
to rate and post a comment
Georgie
re: re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 21, 2013 09:02
You need to check the controlers.Are all of them OK?
Do you know how to check it?
to rate and post a comment
Georgie
re: re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 21, 2013 09:20
something is wrong with the email address
to rate and post a comment
Pedro Zamora
re: re: re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
December 24, 2013 07:27
Dear George,
Let me tell you that, at the moment to check small display, it does not show anything to help me for finding any info of board reported. But actually, when system start with WARM-UP operation send following code sync_lost_ rs and stop rotation of gantry.
Waiting your comments
Thanks
to rate and post a comment
Valentin Cardona
re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
April 17, 2014 01:15
Hi this is Valentin Cardona
My Plus 4 expert give error MAS_700013 when init the warm up
I dopnt have the technical information about errors
Can you help me?
Thanks
to rate and post a comment
Valentin Cardona
re: re: ERRORS ON CT SCANNER SOMATOM PLUS 4 EXPERT
April 17, 2014 01:15
Mu e mail is valentinimagen@hotmail.com