Eureka Linear IV collimator

Tags: Radiology - R/F, EUREKA

(2681)
(673)
(21)

Lenny Place

Eureka Linear IV collimator

March 08, 2017 09:09

I cannot locate the code to perform the auto calibration can someone help me.
Thank you
**redacted**

1 person liked this
Log inor Register

to rate and post a comment

(22)
(4)

Billy-Mac

re: Eureka Linear IV collimator

March 10, 2017 04:46

The security code is the following:
LAST SIZE, MAG SIZE, ENT/SET,
CAL followed by ENT/SET to enter
the code.

Make sure you have 14x17 and 8x10 cassettes and the collimation size sheet before starting.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: Eureka Linear IV collimator

March 12, 2017 10:24

Thank you