Hi, an Achieva 1.5T (R3, F2000) has the same helium level reading (66%) by aprox 3 months (measuring at Console and at RMMU side). No cryo service have been done recently, magnet pressure is normal (aprox 70 mbar) . According to Logs the Boil-Off has lowered in that time frame from 0.038 to 0.026 . Any clue about this?
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
JhonW
F2000 - Helium Level reading does not change
October 14, 2019 10:15
Hi, an Achieva 1.5T (R3, F2000) has the same helium level reading (66%) by aprox 3 months (measuring at Console and at RMMU side). No cryo service have been done recently, magnet pressure is normal (aprox 70 mbar) . According to Logs the Boil-Off has lowered in that time frame from 0.038 to 0.026 . Any clue about this?