kco
FAIL LOGIQ 400
July 03, 2013 01:05
I have a problem with a logiq 400, to start the computer will freeze the image with the LG logo, and program progress bars remain frozen, searched the hard drive and it seems okay, to install a message again error:
SYS ERROR: ERROR SCSI. DEVICE: 0.0 CODE: 20166
CAN SOMEONE HELP ME, THANK YOU.
to rate and post a comment
SHYJU MON
re: FAIL LOGIQ 400
July 04, 2013 08:16
Check your hard disk, scassy card an scassy cable, if all are ok have to repair CPU board, contact shyjumon@hotmail.com
to rate and post a comment
Saqueeb Usmani
re: FAIL LOGIQ 400
July 04, 2013 08:17
change the cpu board this is the cpu board problem
to rate and post a comment
alejandro mendoza
re: FAIL LOGIQ 400
July 04, 2013 08:17
reload software
regards
to rate and post a comment
Lakshmi Narayana
re: FAIL LOGIQ 400
July 04, 2013 08:40
Check the Hard Disk and MVME board.
to rate and post a comment
BINU ACHARI
re: re: FAIL LOGIQ 400
September 23, 2013 11:54
check the CPU the Board
to rate and post a comment
BINU ACHARI
re: FAIL LOGIQ 400
September 23, 2013 11:54
check the CPU board
to rate and post a comment