2.15.1 Re-exposure processing (A61-A63)If error E35 (Insufficient tube voltage), E36 (Tube over voltage), or E37 (Discharge detected) is detected, re-exposure processing will be performed up to two times. In re-exposure processing, the values of kV and mAs are the same as normal exposure processing (see the following figure).
E35
Insufficient tube voltage{MD:2.15.4}
No
- Make sure that CNKV is attached to the CPU board.- Make sure that there is no disconnection or ground fault in the tube voltage detection circuit.- Check the battery voltage during radiography.
E36
Tube over voltage{MD:2.15.3}
No
CPU board3E1-7,3E2-7: L
- Make sure that there is high-voltage discharge at the tube head. -> Check the waveforms between KV+ and KV+G (GND) and between KV- and KV-G (GND) on the CPU board.- Check whether the adjustment value of E36 is not changed to less value.
E37
Discharge detected{MD:2.15.7}
No
CPU board2D3-7: L
- Make sure that there is high-voltage discharge at the tube head.- Make sure that the main transformer is not layer shorted.- Make sure that +kV and -kV are well balanced.
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
xtekvet
FCRGO A61 error
November 06, 2018 10:17
FcrGo has an A61 error message. Can still take images. Manual does not give reason for error or how to fix it. Any suggestions?
to rate and post a comment
Alfred M. Brown
re: FCRGO A61 error
November 07, 2018 08:19
2.15.1 Re-exposure processing (A61-A63)If error E35 (Insufficient tube voltage), E36 (Tube over voltage), or E37 (Discharge detected) is detected, re-exposure processing will be performed up to two times. In re-exposure processing, the values of kV and mAs are the same as normal exposure processing (see the following figure).
to rate and post a comment
Alfred M. Brown
re: FCRGO A61 error
November 07, 2018 08:19
E35
Insufficient tube voltage{MD:2.15.4}
No
- Make sure that CNKV is attached to the CPU board.- Make sure that there is no disconnection or ground fault in the tube voltage detection circuit.- Check the battery voltage during radiography.
E36
Tube over voltage{MD:2.15.3}
No
CPU board3E1-7,3E2-7: L
- Make sure that there is high-voltage discharge at the tube head. -> Check the waveforms between KV+ and KV+G (GND) and between KV- and KV-G (GND) on the CPU board.- Check whether the adjustment value of E36 is not changed to less value.
E37
Discharge detected{MD:2.15.7}
No
CPU board2D3-7: L
- Make sure that there is high-voltage discharge at the tube head.- Make sure that the main transformer is not layer shorted.- Make sure that +kV and -kV are well balanced.