i have a picker Elite 1500 table with a fischer x ray generator. when i command fluoroscopy the console reports an error: o5, o6 or o7, alternatively. exposures doesn´t report any error, but no radiation. any help with this code?
I don't know what generator you have; but if it is a 425/525/or 625 then E5 is a UART Parity error - check console link; E6 and E7 are DC supply voltage too high or too low, respectively. Carefully check E1A1C1-C4 capacitor bank should be roughly 340Vdc +- 10%. Hope it helps.
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
gustavo fiochi
fischer imaging generator
November 13, 2011 06:13
i have a picker Elite 1500 table with a fischer x ray generator. when i command fluoroscopy the console reports an error: o5, o6 or o7, alternatively. exposures doesn´t report any error, but no radiation. any help with this code?
to rate and post a comment
Doug Gray
Re: fischer imaging generator
November 17, 2011 11:45
What generator is it?
to rate and post a comment
gustavo fiochi
Re: fischer imaging generator
November 19, 2011 01:00
i believe is an mx30/50 hf
to rate and post a comment
RICH SEIDEMAN
Re: fischer imaging generator
November 19, 2011 10:38
I don't know what generator you have; but if it is a 425/525/or 625 then E5 is a UART Parity error - check console link; E6 and E7 are DC supply voltage too high or too low, respectively. Carefully check E1A1C1-C4 capacitor bank should be roughly 340Vdc +- 10%. Hope it helps.
Rich Seideman
to rate and post a comment