fisher srx 30 hf problem

March 23, 2010 08:00

hello everybody i have a problem with a high frecuency generator fisher imaging model srx30hf someone have experience with this equiment? any help will be apprecciate

Log inor Register

to rate and post a comment

Doug Gray

Re: fisher srx 30 hf problem

March 26, 2010 11:47

Can you send a pictureof the control?
doug@allstarxray.com

Log inor Register

to rate and post a comment

Doug Gray

Re: fisher srx 30 hf problem

April 01, 2010 02:52

Manuel,

I had a similar problem, the unit would flash all digits and shut off. There was a cold solder joint in the control console on the power supply transformer that is mounted on the board. There is a dip switch SW2-1 that initializes the EEROM and clears Cal. Data. The switch is in the generator (controller board) not the console.

Log inor Register

to rate and post a comment

(120)
(6)
(1)

manuel valdez

Re: fisher srx 30 hf problem

April 05, 2010 01:39

hello Doug i changed the console and the system is not working
i need clear the data cal??
thanks

Log inor Register

to rate and post a comment