hello everybody i have a problem with a high frecuency generator fisher imaging model srx30hf someone have experience with this equiment? any help will be apprecciate
I had a similar problem, the unit would flash all digits and shut off. There was a cold solder joint in the control console on the power supply transformer that is mounted on the board. There is a dip switch SW2-1 that initializes the EEROM and clears Cal. Data. The switch is in the generator (controller board) not the console.
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
manuel valdez
fisher srx 30 hf problem
March 23, 2010 08:00
hello everybody i have a problem with a high frecuency generator fisher imaging model srx30hf someone have experience with this equiment? any help will be apprecciate
to rate and post a comment
Doug Gray
Re: fisher srx 30 hf problem
March 26, 2010 11:47
Can you send a pictureof the control?
doug@allstarxray.com
to rate and post a comment
Doug Gray
Re: fisher srx 30 hf problem
April 01, 2010 02:52
Manuel,
I had a similar problem, the unit would flash all digits and shut off. There was a cold solder joint in the control console on the power supply transformer that is mounted on the board. There is a dip switch SW2-1 that initializes the EEROM and clears Cal. Data. The switch is in the generator (controller board) not the console.
to rate and post a comment
manuel valdez
Re: fisher srx 30 hf problem
April 05, 2010 01:39
hello Doug i changed the console and the system is not working
i need clear the data cal??
thanks
to rate and post a comment