32 50 2 kVp COMMUTATION FAIL HARDWARE ERROR
Unable to clear kV commutation fail latch. Checked at the start of exposure
sequence. Most probable cause is the opening of SCR Fuse F9 (on units that
have them). See “KV Commutation Failure Detection” on page 81.
KV Commutation Failure Detection
Circuit
Description
Circuit Function: This Circuit monitors the Inverter Power supply voltage.
If it drops to (near) zero Volts it performs the following functions:
• disables the KV Alternating Pulse Circuit (see page 56)
• disables the KV Gate Drive Circuits (see page 57)
• signals the RPU mprocessor (which terminates the exposure and sends a
flag to the OC to display the error)
Circuit Flow: Two thin wires, red (+525VDC) and white (RETURN), connect
from (across) the Inverter Power Supply to TB5-1 and TB5-2 of the Gate Drive
PCB. The voltage of the Inverter Power Supply is divided by R30, R31, L6 and
the LED portion of IC7. If the inverter cross fires (page 58) the voltage drops to
zero volts. As it approaches zero volts L6 and the LED portion of opto-coupler
IC7 no longer illuminate. This causes the output of IC7 (pin 7) to go low. This
turns off the LED portion of opto-couplerIC9 and causes its output to go low.
This low signal goes two places:
1. Off the board as -KV COMM FAIL at J1-13. This signal moves up the Gate
Drive Ribbon Cable and across the Backplane PCB into the KV Control
PCB P6-21 where it is called KVCF. It next passes through two inverters to
Flip-Flop IC12B J-K driving its “Preset” input low causing its Q output to go
high and its Q output to go low. These outputs effectively split the signal
and function as follows:
• The Q output is inverted by IC10F and then passed through EKV9 to
hold low the J-K inputs of IC12A. This disables the Alternating Pulse
Circuit (see page 56).
• The Q output becomes COM_FAIL and is sent to IC5 through EKV0.
IC5 signals the mprocessor to terminate the exposure and send a
flag to the OC.
2. To one input of IC13B causing a low at its output. This triggers the input of
one-shot IC11B and produces a 0.1 second, positive going pulse at its output
(pin-5). This pulse is passed to IC13A and IC13C and prevents them
from passing the low going inverter drive pulses, preventing continued
operation of the inverter.
Hi Lenny, I'm installing a Trex TM50, this is not responding to calibration of high mA 600mA, up to 200mA, the mA curve is fairly flat, above this value the curve falls to the point that the equipment crashes Low current, Error Flag 7 .
I have checked the PCB's Isolation feedback mA, mA drive control and gate drive. Are apparently well, I have a manual 9026.251 Revised 9890 that is not 100% compatible with the unit, I have no manual for the table or the wall bucky.
I would greatly appreciate if you can help me.
The models of the system are:
TM50 6626,235 S / N 968816,
Table 6807.310
Wall Bucky 6616.205
OC S / W see 4.8.0 1.0.2 comp.
Teck see 1.0.5 Memory 1.0.2 comp
SPS S / W see 4.8.0 4/25/96
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
ProfessionalMedSalesLLC
Flag 32 error in conmutation HV
January 28, 2014 05:33
we have a X-ray Hologic trex TM50
in the console, this error appears:
Flag 32 error in conmutacion High Voltage
someone can help me to know it is and how to fix it?
Thanks
to rate and post a comment
Lenny Place
re: Flag 32 error in conmutation HV
January 29, 2014 06:13
32 50 2 kVp COMMUTATION FAIL HARDWARE ERROR
Unable to clear kV commutation fail latch. Checked at the start of exposure
sequence. Most probable cause is the opening of SCR Fuse F9 (on units that
have them). See “KV Commutation Failure Detection” on page 81.
KV Commutation Failure Detection
Circuit
Description
Circuit Function: This Circuit monitors the Inverter Power supply voltage.
If it drops to (near) zero Volts it performs the following functions:
• disables the KV Alternating Pulse Circuit (see page 56)
• disables the KV Gate Drive Circuits (see page 57)
• signals the RPU mprocessor (which terminates the exposure and sends a
flag to the OC to display the error)
Circuit Flow: Two thin wires, red (+525VDC) and white (RETURN), connect
from (across) the Inverter Power Supply to TB5-1 and TB5-2 of the Gate Drive
PCB. The voltage of the Inverter Power Supply is divided by R30, R31, L6 and
the LED portion of IC7. If the inverter cross fires (page 58) the voltage drops to
zero volts. As it approaches zero volts L6 and the LED portion of opto-coupler
IC7 no longer illuminate. This causes the output of IC7 (pin 7) to go low. This
turns off the LED portion of opto-couplerIC9 and causes its output to go low.
This low signal goes two places:
1. Off the board as -KV COMM FAIL at J1-13. This signal moves up the Gate
Drive Ribbon Cable and across the Backplane PCB into the KV Control
PCB P6-21 where it is called KVCF. It next passes through two inverters to
Flip-Flop IC12B J-K driving its “Preset” input low causing its Q output to go
high and its Q output to go low. These outputs effectively split the signal
and function as follows:
• The Q output is inverted by IC10F and then passed through EKV9 to
hold low the J-K inputs of IC12A. This disables the Alternating Pulse
Circuit (see page 56).
• The Q output becomes COM_FAIL and is sent to IC5 through EKV0.
IC5 signals the mprocessor to terminate the exposure and send a
flag to the OC.
2. To one input of IC13B causing a low at its output. This triggers the input of
one-shot IC11B and produces a 0.1 second, positive going pulse at its output
(pin-5). This pulse is passed to IC13A and IC13C and prevents them
from passing the low going inverter drive pulses, preventing continued
operation of the inverter.
to rate and post a comment
ProfessionalMedSalesLLC
re: re: Flag 32 error in conmutation HV
January 29, 2014 07:41
Thnak you for you help. I'll do what you tell me and I mention anything.
to rate and post a comment
Lenny Place
re: re: re: Flag 32 error in conmutation HV
January 30, 2014 02:30
do you need the service documents on this unit
to rate and post a comment
ProfessionalMedSalesLLC
re: re: re: re: Flag 32 error in conmutation HV
January 30, 2014 03:58
if you have and want share with me ....will be fantastic!! professionalmedsales@gmail.com
thanks
to rate and post a comment
Lenny Place
re: re: re: re: re: Flag 32 error in conmutation HV
January 30, 2014 04:32
look for a email from sendthisfile
to rate and post a comment
ProfessionalMedSalesLLC
re: re: re: re: re: re: Flag 32 error in conmutation HV
January 31, 2014 06:32
Thnk you so much!
to rate and post a comment
Alfonso Nino
re: re: re: re: Flag 32 error in conmutation HV
April 18, 2014 07:26
Hi Lenny, I'm installing a Trex TM50, this is not responding to calibration of high mA 600mA, up to 200mA, the mA curve is fairly flat, above this value the curve falls to the point that the equipment crashes Low current, Error Flag 7 .
I have checked the PCB's Isolation feedback mA, mA drive control and gate drive. Are apparently well, I have a manual 9026.251 Revised 9890 that is not 100% compatible with the unit, I have no manual for the table or the wall bucky.
I would greatly appreciate if you can help me.
The models of the system are:
TM50 6626,235 S / N 968816,
Table 6807.310
Wall Bucky 6616.205
OC S / W see 4.8.0 1.0.2 comp.
Teck see 1.0.5 Memory 1.0.2 comp
SPS S / W see 4.8.0 4/25/96