I have fluoroscan premier with discharge battery in main board. When I replace clock and memory chip contain this battery - windows NT start up and premier program load. In diagnostic screen I have password field. Passwords are type in file on disc - Checkmate for admin and user for user. But when I log in and confirm patient data , program not respond. I think its a CMOS settings. I try this HDD on two units with similar results. COM PORT 1 and 2 work - one for terminal (PASWORD FIELD ETC.) and second for camera rotation. I think its may by LPT adres , IRQ or another settings or software damaged.
Lenny Place - thanks for this service note, but I have oldest main board. It's have AMIBIOS not PHOENIX-AWARD.
If I set LPT to 378/IRQ3 ECP DMA mode 3 system stop permanently. Ctrl Alt Del not work. Its look at IRQ or address problem.
Eryk,
i/m working on a encore premier, change the 3v. battery , computer says, "CMOS checksum error- Defaults loaded", so i went to cmos bios ( rev 1.30 2002) default and load it, , the system came up, nice, but, unresponsive to any key on the keyboard, and tochscreen. " could this be a problem on the c mos memory, or software?. anyone have come across this problem?
if is a software problem, does any one have the procedure to load the software?
regards.
for any help on the matter.
george
Lightly used system available direct from the end user customer. MONARCH QUEST
Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
Eryk Michałowski
Fluoroscan Premier problem
March 06, 2014 02:25
I have fluoroscan premier with discharge battery in main board. When I replace clock and memory chip contain this battery - windows NT start up and premier program load. In diagnostic screen I have password field. Passwords are type in file on disc - Checkmate for admin and user for user. But when I log in and confirm patient data , program not respond. I think its a CMOS settings. I try this HDD on two units with similar results. COM PORT 1 and 2 work - one for terminal (PASWORD FIELD ETC.) and second for camera rotation. I think its may by LPT adres , IRQ or another settings or software damaged.
to rate and post a comment
Lenny Place
re: Fluoroscan Premier problem
March 06, 2014 09:04
please send me a email address I can send you a service note reference this
to rate and post a comment
Eryk Michałowski
re: re: Fluoroscan Premier problem
March 10, 2014 08:33
Hi
Can You send this manual to eryk5_@_op.pl (remove all "_" from address).
to rate and post a comment
Eryk Michałowski
re: re: Fluoroscan Premier problem
March 10, 2014 08:33
Hi
Can You send this manual to eryk5_@_op.pl (remove all "_" from address).
to rate and post a comment
shih-chung chiu
re: re: Fluoroscan Premier problem
March 10, 2014 03:39
hi
I havd same problem ,can you send to me thank's
email: andy_chiu@seed.net.tw
to rate and post a comment
MedsystemsPoland
re: re: Fluoroscan Premier problem
March 10, 2014 06:13
Good Morning,
Please send solution on
Info@aukcjamedyczna.pl
Regards
Rafal
to rate and post a comment
Gerardo Fuentes
re: re: Fluoroscan Premier problem
March 11, 2014 08:06
Hi Lenny I have several problems with same system can you email me the service manual ? to g_fuentesg@hotmail.com ....best regards.
to rate and post a comment
Eryk Michałowski
re: re: re: Fluoroscan Premier problem
March 12, 2014 08:50
Lenny Place - thanks for this service note, but I have oldest main board. It's have AMIBIOS not PHOENIX-AWARD.
If I set LPT to 378/IRQ3 ECP DMA mode 3 system stop permanently. Ctrl Alt Del not work. Its look at IRQ or address problem.
to rate and post a comment
Miguel Chavira
re: re: re: re: Fluoroscan Premier problem
May 19, 2015 09:00
Hi! can you solve the problema, because i have the same problema.
to rate and post a comment
bashir ahmad
re: re: Fluoroscan Premier problem
October 10, 2015 09:59
Please if you have its software and service manual i need it my email is
to rate and post a comment
bashir ahmad
re: re: re: Fluoroscan Premier problem
October 12, 2015 09:24
My email is bashir30474@yahoo.com
to rate and post a comment
IngJose
re: re: Fluoroscan Premier problem
November 16, 2022 06:37
hi lenny I have a equipment and is not working show "system start up please wait" in principal monitor could you help me?
to rate and post a comment
Miguel Chavira
re: Fluoroscan Premier problem
May 19, 2015 09:00
Hi! Eryk, can you solve yor problema? i have a similar trouble.
to rate and post a comment
Moshe Alkalay
re: Fluoroscan Premier problem
June 23, 2015 03:20
call fluoroscan- they will send u info to set up BIOS
u can call me - I have some but not sure it will work for u
sales@htig.com
to rate and post a comment
Moshe Alkalay
re: re: Fluoroscan Premier problem
October 12, 2015 03:49
now it boots up but will not produce x ray.... light for x ray comes on, but NO IMAGE...
to rate and post a comment
Miguel Chavira
re: re: Fluoroscan Premier problem
October 12, 2015 06:28
do you have the BIOS set up?
can you send me the information please?
asbsadecv@Hotmail.com
to rate and post a comment
OK
re: Fluoroscan Premier problem
March 11, 2021 10:26
Eryk,
i/m working on a encore premier, change the 3v. battery , computer says, "CMOS checksum error- Defaults loaded", so i went to cmos bios ( rev 1.30 2002) default and load it, , the system came up, nice, but, unresponsive to any key on the keyboard, and tochscreen. " could this be a problem on the c mos memory, or software?. anyone have come across this problem?
if is a software problem, does any one have the procedure to load the software?
regards.
for any help on the matter.
george
to rate and post a comment
Moshe Alkalay
re: re: Fluoroscan Premier problem
March 12, 2021 10:39
connect another keyboard directly at base and see if it works. should be connection
to rate and post a comment