fresenius 2008H problem

Tags: Dialysis, FRESENIUS

(33)
(3)

Juan M Maciel

fresenius 2008H problem

July 16, 2015 02:33

I have a fresenius 2008H with this problem, when dyalisis mode is run after 10 or 15 seconds stop and appear AIR legend in display.

Somebody can help me??

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: fresenius 2008H problem

July 16, 2015 06:19

you could have an air leak. have you performed the Induced Positive Tests?

Log inor Register

to rate and post a comment

(33)
(3)

Juan M Maciel

re: re: fresenius 2008H problem

July 21, 2015 06:09

air problem already fix it, I have problem with low conductivity (about 12.6). I change sensor board, need calibrate conductivity??

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: fresenius 2008H problem

July 21, 2015 06:31

do you need the service manuals? if so send me an email

Log inor Register

to rate and post a comment

(33)
(3)

Juan M Maciel

re: re: re: re: fresenius 2008H problem

July 22, 2015 08:26

Already fix this problem, but need the manuals. Please give an email address.

Log inor Register

to rate and post a comment

(33)
(3)

Juan M Maciel

re: re: fresenius 2008H problem

July 22, 2015 08:27

The problem was solved, thank's. Sensor board was damaged.

Log inor Register

to rate and post a comment