Hello, I am looking to see if other people have Fuji DR and experience similar problems that we have. I Ghost Image sometimes appear multiple images later. A Brief explanation was provided by our service rep. He indicated you have two image collection types a hard x-ray or soft x-ray. So an example of the Ghost occurring is going from a patient that was a hard x-ray described as hot dose to a soft low dose like extremity and you see the prior study like chest.
We all so have lines running the opposite direction the grid lines would appear in.
If you have experienced these problems do you have a resolution or better understanding of why these occur?
re-do your lag calibration, but use 150+ mr instead of 100 as the manual suggests. Known issue with the D-evo cal. Lines may indicate that the panel was dropped.
Redo Lag cal as mentioned. Also don't use during emergency use mode.... ever. And most importantly upgrade the MC software to the latest. Latest versions have fix for ghosting
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Mark Parris
FUJI DR
April 01, 2014 04:44
Hello, I am looking to see if other people have Fuji DR and experience similar problems that we have. I Ghost Image sometimes appear multiple images later. A Brief explanation was provided by our service rep. He indicated you have two image collection types a hard x-ray or soft x-ray. So an example of the Ghost occurring is going from a patient that was a hard x-ray described as hot dose to a soft low dose like extremity and you see the prior study like chest.
We all so have lines running the opposite direction the grid lines would appear in.
If you have experienced these problems do you have a resolution or better understanding of why these occur?
to rate and post a comment
Ray Kates
re: FUJI DR
April 03, 2014 11:07
Re-do your lag calibration, but bump your dose up to 150+ mr. This is a known issue with Fuji and this should take care of the problem.
to rate and post a comment
Ray Kates
re: FUJI DR
April 03, 2014 11:07
re-do your lag calibration, but use 150+ mr instead of 100 as the manual suggests. Known issue with the D-evo cal. Lines may indicate that the panel was dropped.
to rate and post a comment
Mustafa RUZGAR
re: FUJI DR
April 04, 2014 08:16
PLEASE! Agfa medical devices and related services in your hand all my e-mail address to send documents musti_1988_42@hotmail.com
to rate and post a comment
Kevin Dietz
re: FUJI DR
March 19, 2015 01:40
Redo Lag cal as mentioned. Also don't use during emergency use mode.... ever. And most importantly upgrade the MC software to the latest. Latest versions have fix for ghosting
to rate and post a comment