Over 150 Missouri Auctions End Today - Bid Now

GANTRY ACQUISITION SYSTEM CANNOT COMPLY

Tags: CT

(13)
(4)

ALVARO BALAREZO

GANTRY ACQUISITION SYSTEM CANNOT COMPLY

May 29, 2013 05:23

Hi Forum:
During the procedure, get the next message:
GANTRY ACQUISITION SYSTEM CANNOT COMPLY, PLEASE RETRY.
can someone help me with this
Thank you in advance
Alvaro

1 person liked this
Log inor Register

to rate and post a comment

(23)
(2)

Ed Perez

re: GANTRY ACQUISITION SYSTEM CANNOT COMPLY

May 29, 2013 09:43

Gotta give more info such as !! What kind of unit Sounds Siemens or Marconi

Log inor Register

to rate and post a comment

(147)
(22)

Wang Zhixuan

re: GANTRY ACQUISITION SYSTEM CANNOT COMPLY

May 29, 2013 09:43

What is you said equipment? What is its model?You must tell
that to us.

Log inor Register

to rate and post a comment

(13)
(4)

ALVARO BALAREZO

re: GANTRY ACQUISITION SYSTEM CANNOT COMPLY

May 29, 2013 09:43

The system is MX8000 DUAL VEXP PHILIPS
Alvaro

Log inor Register

to rate and post a comment

(23)
(2)

Ed Perez

re: re: GANTRY ACQUISITION SYSTEM CANNOT COMPLY

May 29, 2013 11:40

pow look at the system error log Gantry can't comply because some one (MCB's) did not return the hand shake... If you look at the error log click on the error code several numbers in sequence there it will tell ya the problemo If not check The MCU and Fiber optic's Make sure the raid disk drive are all operational

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)