Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

GE 2000 problem

(17)
(1)

Karen Harvey

GE 2000 problem

February 07, 2018 07:25

I faced a problem in cathlab GE innova 2000.
The problem that a customer called me facing a problem that the system is not moving when I reached there I found the CPU borad (Motorola)of the positioner is defected(red hault led is on)
I replaced it with a working one the CPU worked well.
But the c arc move then stop moving suddenly and the table is not moving up or down.
I'm not able to calibrate the whole system as the table didn't move up or down and I don't have the LC backup software for this system.
I think the only way to solve this problem is to deactivate the collision risks but I don't know how to do it.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

OEM Philips & Siemens parts with Warranty!

Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com

stats

Upcoming Auction Events

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)