GE 2000 technical issue

(17)
(1)

Karen Harvey

GE 2000 technical issue

February 07, 2018 07:26

I faced a problem in cathlab GE innova 2000.
The problem that a customer called me facing a problem that the system is not moving when I reached there I found the CPU borad (Motorola)of the positioner is defected(red hault led is on)
I replaced it with a working one the CPU worked well.
But the c arc move then stop moving suddenly and the table is not moving up or down.
I'm not able to calibrate the whole system as the table didn't move up or down and I don't have the LC backup software for this system.
I think the only way to solve this problem is to deactivate the collision risks but I don't know how to do it.

Log inor Register

to rate and post a comment

(78)
(7)

R GOPAL

re: GE 2000 technical issue

February 08, 2018 10:08

Hi Karen Harvey,

At which angle did ARC stop? Is there any collision?
If collision is not there, I sincere feeling is there are two problems. One with the ARC movement and another with the TABLE. Try to address the problems separately.

If you have any BACKUP floppy for the 'C'ARM try loading it at first.

If not, You can connect your LAPTOP to the machine and Calibrate the 'C'Arm.

All collision switches will be automatically deactivated when you do the calibration.

By the which is your location?

You may write to me at aparnaelectromech@gmail.com

R.Gopal

1 person liked this