Karen Harvey
GE 3100 problem
October 23, 2018 09:03
I faced an intermittent problem: collimator power failure in cathlab GE innova 3100.
This error appear from time to time.
When I run the diagnostic from SUIF it states that the collimator is good but when I run it when the problem exist it didn't pass.
I checked the voltage to collimator it's ok.
I tried to reload the RTAC software but with no success.
Any advise?
to rate and post a comment
Wang Zhixuan
re: GE 3100 problem
October 24, 2018 08:41
Have you tried to change the hardware?
to rate and post a comment
Karen Harvey
re: re: GE 3100 problem
October 25, 2018 03:15
What do youean by changing the hardware??
to rate and post a comment
Aleksander Proof
re: GE 3100 problem
October 24, 2018 08:41
Hi!
Check the curtains on the collimator, often they cling to each other
to rate and post a comment
Karen Harvey
re: re: GE 3100 problem
October 25, 2018 03:15
Can you explain more
to rate and post a comment