Karen Harvey
GE 3100 problem
October 23, 2018 09:06
I faced an intermittent problem: collimator power failure in cathlab GE innova 3100.
This error appear from time to time.
When I run the diagnostic from SUIF it states that the collimator is good but when I run it when the problem exist it didn't pass.
I checked the voltage to collimator it's ok.
I tried to reload the RTAC software but with no success.
Any advise?
to rate and post a comment
Medical Support
re: GE 3100 problem
October 26, 2018 01:53
Have you checked collimator power on mother regulation board when machine is in problem phase?
to rate and post a comment
Karen Harvey
re: re: GE 3100 problem
October 27, 2018 10:27
Yes it is ok
to rate and post a comment
Karen Harvey
re: re: GE 3100 problem
October 27, 2018 10:28
When the error happens I opened the blades are totally closed.
When I try to move the blades by the joystick it didn't respond.
When I restart it it works normally for a while then it happened again that the blades are totally closed
to rate and post a comment
MRAdmin
re: re: re: GE 3100 problem
October 29, 2018 12:56
As I told you below, you need to clean the movable parts inside the collimator
to rate and post a comment
MRAdmin
re: GE 3100 problem
October 27, 2018 10:26
Open the covers and try to clean the blades with WD-40 or any think like it.
Some time a sticky blades are the problem, also when you have this problem, open the cover and have a look inside.
to rate and post a comment
Peter Azmy
re: re: GE 3100 problem
November 29, 2018 09:43
Did it work ?
to rate and post a comment
MRAdmin
re: re: re: GE 3100 problem
November 30, 2018 09:39
Do you have the same problem?
to rate and post a comment