Pablo Piccinini

GE 600T /CGR E160/150

December 22, 2015 10:51

Hi colleagues, I have a problem with a mammo unit GE/CGR 600T.
Explanation: when I switch it on, after a long time in stand-by, they work properly until It display error E10, I switch it OFF and ON, after that two errors are shown: E150 / E160 again I switch OFF an ON but two errors remain and inhibit the X Ray exposure.
I check 400PL1, 400PL3 and 400PL3 they have no problems,
two filaments are OK,
HV cable are OK.

Any idea, TIP o suggestion???

Thanks in advance.

Pablo.

Log inor Register

to rate and post a comment

(277)
(41)

Luis Fernando Garcia Sanchez

re: GE 600T /CGR E160/150

December 22, 2015 04:07

Error 10 generator control fault. Error 150 filament current fault .1 . Error 160 filament .3 .

Log inor Register

to rate and post a comment

(88)
(4)

Edurado Madeiros

re: GE 600T /CGR E160/150

December 23, 2015 08:59

Check hv/mesure bord front HV tank.
Some time bad connection or fault diodos is

Log inor Register

to rate and post a comment

(88)
(4)

Edurado Madeiros

re: GE 600T /CGR E160/150

December 23, 2015 08:59

Check kv/ma mesure board. May be diode broken or short. If not check Wiring.
150 and 160 i means all focus is open

Log inor Register

to rate and post a comment