Hi colleagues, I have a problem with a mammo unit GE/CGR 600T.
Explanation: when I switch it on, after a long time in stand-by, they work properly until It display error E10, I switch it OFF and ON, after that two errors are shown: E150 / E160 again I switch OFF an ON but two errors remain and inhibit the X Ray exposure.
I check 400PL1, 400PL3 and 400PL3 they have no problems,
two filaments are OK,
HV cable are OK.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Pablo Piccinini
GE 600T /CGR E160/150
December 22, 2015 10:51
Hi colleagues, I have a problem with a mammo unit GE/CGR 600T.
Explanation: when I switch it on, after a long time in stand-by, they work properly until It display error E10, I switch it OFF and ON, after that two errors are shown: E150 / E160 again I switch OFF an ON but two errors remain and inhibit the X Ray exposure.
I check 400PL1, 400PL3 and 400PL3 they have no problems,
two filaments are OK,
HV cable are OK.
Any idea, TIP o suggestion???
Thanks in advance.
Pablo.
to rate and post a comment
Luis Fernando Garcia Sanchez
re: GE 600T /CGR E160/150
December 22, 2015 04:07
Error 10 generator control fault. Error 150 filament current fault .1 . Error 160 filament .3 .
to rate and post a comment
Edurado Madeiros
re: GE 600T /CGR E160/150
December 23, 2015 08:59
Check hv/mesure bord front HV tank.
Some time bad connection or fault diodos is
to rate and post a comment
Edurado Madeiros
re: GE 600T /CGR E160/150
December 23, 2015 08:59
Check kv/ma mesure board. May be diode broken or short. If not check Wiring.
150 and 160 i means all focus is open
to rate and post a comment