Hi everyone, I have a 800T senographe with error E06
068/025 floating calcul. Address: 220EBC.
When I turn on the mammo passes the test well but after 5 min the error comes out.
Some of you that can help me to repair this fault
Do a backup of the calibrations and perform a full calibration of the system. Check also the NVram battery status. The E06 error comes from corrupt/bad calibration data.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Julian Aguilera
Ge 800T E06
May 09, 2017 10:08
Hi everyone, I have a 800T senographe with error E06
068/025 floating calcul. Address: 220EBC.
When I turn on the mammo passes the test well but after 5 min the error comes out.
Some of you that can help me to repair this fault
thank you very much for your help
to rate and post a comment
Hugo Macias
re: Ge 800T E06
May 10, 2017 09:57
Do a backup of the calibrations and perform a full calibration of the system. Check also the NVram battery status. The E06 error comes from corrupt/bad calibration data.
to rate and post a comment
Tecsila Service
re: Ge 800T E06
May 17, 2017 10:28
You need to calibrate AEC.
Did your machine is working in Normal KV/mA mode?
to rate and post a comment