hello can anyone assist me to fix the E64 162/11 in one Ge 800T? im not familiar with this system, i have the service manual but there is nothing explaining this problem. thanks
to rate and post a comment
Staate Hayward
Re: GE 800T E64 Error 162/11
March 12, 2010 04:47
Filter wheel was moved but it didn't see a flag go through the opto coupler. Loose connector or dust or motor drive not working correctly. Check filter wheel inside the tube covers.