Galen Hiveley
GE 800T error code E13
January 18, 2016 09:37
I need some help. 2002 DOM GE 800T, the original console intermittently tries to reset. When replacing the same part number console (DOM 2001)it boots to an error code E13. Is there a compatibility issue with the difference of the DOM's or does this error code point to something else?
Thanks,
Galen
to rate and post a comment
Lenny Place
re: GE 800T error code E13
January 19, 2016 06:06
did you restore the backups?
to rate and post a comment
Galen Hiveley
re: re: GE 800T error code E13
January 20, 2016 08:26
I was wrong on the error, it is E12, cksum error. DOM 2002. The service manual says to change the setting of switch 8 of B1 on the CPU board, then power up and the double error key to get into the install menu to get to the cksum reset menu. Switch 8? is that in the console CPU or is that a board within the generator cabinet?
Thanks,
Galen