JIFFY PAUL AKKARAKARAN

GE Advantx tilt c positioner PRD error code

March 15, 2011 05:38

Dear Colleagues! I am having a problem for Ge Advantx tilt c positioner system PRD error. The CPU board's 2 digit error display shows 50 and then shows 56 on the power on. This error started after the table hit the x-ray tube head in an accident. What does this mean? I could get error codes up to 46 and then from 60 onwards. But I could not find the 50 to 59 in the diagnostics manual. If anyone knows what this means your help is very much solicited

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE Advantx tilt c positioner PRD error code

March 16, 2011 04:22

PRD of the Tilt-C CPU bd. or the PRD of the Advantx-E (or Advantx-1) CPU bd??? Once the collision had occurred,the only movement allowed should be reverse or out of the collision. If only some movements are allowed you may need to turn off the limits to allow any and all movement, but this will allow you to get into worse situation as well. Verify all collision sensors have been released (two spring and cable activated switches in the tube port cover).
shayward@sodexoctm.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)