Over 500 Total Lots Up For Auction at One Location - NJ 06/15

ge alpha rt mammography error

Tags: Mammo, GE

(65)
(7)

ghazy

ge alpha rt mammography error

October 27, 2015 08:26

Hi
I have error (Er 5) in GE alpha rt mammography any help please.

Regards

Log inor Register

to rate and post a comment

(12)
(2)

Manuel Maia

re: ge alpha rt mammography error

October 28, 2015 08:58

5.7 E.05 Error message explanation (software 5.08 and higher)
Error E05 indicates that the preset exposure counter limit has been exceeded.
The unit does not allow further exposures until the limit has been adjusted.
USAGE
- Entering in the limit setting: Press Den- and mAs- all the time and
switch the power on. When display starts blinking, keep pressing Denand
mAs- and reset the program by pressing AutoExp.
- Changing the limit: Press Den- or Den+, 0 means no limit (OFF), other
values indicate amount of the accepted exposures. Altogether there are
50-50 000 possible limit variations.
- Exiting the limit setting: Press kV-.

Best Regards manuelmaia.100@gmail.com

Log inor Register

to rate and post a comment

(88)
(4)

Edurado Madeiros

re: ge alpha rt mammography error

October 28, 2015 08:58

Try checking at first power supply TP2 and TP3 +5v,on C- Arm control board, then change eeprom and last change whole board

Log inor Register

to rate and post a comment

(88)
(4)

Edurado Madeiros

re: ge alpha rt mammography error

October 29, 2015 12:46

Your error is Er 5 or E 05 ?

Log inor Register

to rate and post a comment

(65)
(7)

ghazy

re: re: ge alpha rt mammography error

November 01, 2015 04:44

er5

Log inor Register

to rate and post a comment

(88)
(4)

Edurado Madeiros

re: re: re: ge alpha rt mammography error

November 04, 2015 01:47

give me your email to help you

Log inor Register

to rate and post a comment

(65)
(7)

ghazy

re: re: re: re: ge alpha rt mammography error

November 04, 2015 03:36

emadgazwan@yahoo.com