Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

GE AMX 4 Error Code 102 - Charging

Tags: Radiology - R/F, GE

(332)
(37)
(60)

Gary Benitez

GE AMX 4 Error Code 102 - Charging

June 08, 2022 10:01

Hi Guys,

We are getting an error code of 102 related to a charging error on our GE AMX 4.

Does anyone know what part needs to be changed or re-calibrated?

Please advise.

Log inor Register

to rate and post a comment

(23)
(6)
(1)

Raymundo Moreno

re: GE AMX 4 Error Code 102 - Charging

June 09, 2022 11:40

On charger board, check J11. On male or female connector you may observe bad contact due heathing. Fix it and recalibrate charger board.

Log inor Register

to rate and post a comment

(10)
(1)

Okuwin

re: GE AMX 4 Error Code 102 - Charging

June 09, 2022 11:40

Hello Gary, I would first check your batteries to be sure one of them is not bad. Next, I would try a recalibration of the charging circuit. Finally, if those do not solve your issue replace the charging board. Hope this helps, Bill

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: GE AMX 4 Error Code 102 - Charging

June 10, 2022 09:47

Hello Gary,
Have you checked the plug receptacle and line cord reel?

BATTERY VOLTAGE MUST BE 113 VOLTS OR LESS TO DO LOAD TEST!
DO NOT CHARGE BATTERIES BEFORE LOAD TEST!!

a. Connect Oscilloscope From Circuit Breaker To Ground - Use Sample Mode
b. Make Exposures At 80, 100 KVP And 125 KVP, About 10 MAS
c. Read Battery Voltage And Average The Ripple At The Bottom
d. Battery Voltage Should Drop About 13 VDC +/-10% At 80 KVP
e. Battery Voltage Should Drop About 16 VDC +/-10% At 100KVP
f. Battery Voltage Should Drop About 17 VDC +/-10% At 125KVP
4. If Batteries Pass Load Test, Unit May Require Recalibration Of MAS, KVP,
Taps And Filament Current Tables
5. If Batteries Do Not Pass The Load Test, You Get To Replace Them!

Hope this helps!

Log inor Register

to rate and post a comment

(2)

Marco Antonio Vasquez Cortes

re: GE AMX 4 Error Code 102 - Charging

June 15, 2022 09:17

Hi Gary re-calibrated

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)