GE AMX4 error 43A

(9)
(1)

Carlos Ben

GE AMX4 error 43A

August 27, 2009 02:04

43A is not listed , any information?

Thanks !

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE AMX4 error 43A

August 31, 2009 02:26

Rotor interlock signal failure, Rotor control bd? 60Hz inverter not working (CPU)? Rotor / charger bd? Collimator light operates off of the same inverter.

Log inor Register

to rate and post a comment

(9)
(1)

Carlos Ben

Re: GE AMX4 error 43A

October 19, 2009 12:55

bad rotor control Bd , Thanks a lot !