Rotor interlock signal failure, Rotor control bd? 60Hz inverter not working (CPU)? Rotor / charger bd? Collimator light operates off of the same inverter.
From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729
Carlos Ben
GE AMX4 error 43A
August 27, 2009 02:04
43A is not listed , any information?
Thanks !
to rate and post a comment
Staate Hayward
Re: GE AMX4 error 43A
August 31, 2009 02:26
Rotor interlock signal failure, Rotor control bd? 60Hz inverter not working (CPU)? Rotor / charger bd? Collimator light operates off of the same inverter.
to rate and post a comment
Carlos Ben
Re: GE AMX4 error 43A
October 19, 2009 12:55
bad rotor control Bd , Thanks a lot !