Eric Bem

Ge amx4 plus

March 03, 2026 11:39

Good day brother
Please I need your help on GE AMX 4 plus it is displaying
error 458 and it not capturing images please what can I do to fix it sir
Thanks

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: Ge amx4 plus

March 04, 2026 11:37

ERROR 458 (Could Be With ERROR 320)
1. F93 Blown On Rotor Board
2. Check For Phase 1 & 2 Signals Coming From CPU Board At
Anode Of CR53 And CR66
3. Check For Correct Wiring Of TS2 On Rotor/Charger Panel
AMX1A3
4. Check For Correct Plug Location And Lead Orientation Of
All
Connections To The Rotor Board
5. Check For Correct Wiring Of BLK/WH/GRN Leads To Tube
Stator From J4 Of Rotor Bd.
6. Check For Correct Lead Orientation To CR4 On AMX1A4 Panel
7. Check That R6 Above Autotransformer On AMX1A4 Panel

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: re: Ge amx4 plus

March 04, 2026 03:19

error 320 is rotor, is just another way to narrow it down, but you don't have error 320

Log inor Register

to rate and post a comment

(20)
(2)

Johnson Kaale

re: Ge amx4 plus

March 04, 2026 11:37

Check Rotor Board Fuse: Inspect the 6A fuse (F983) on the Rotor Board

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: Ge amx4 plus

March 04, 2026 11:38

Does this error occur together with error 320?

Error 458

1. F93 blown on rotor board

2.check phase 1&2 signals coming from CPU board at Anode CR53 and CR66

3.check for correct wiring of TS2 on rotor/charge panel AMX1A3

4.check correct plug location and lead orientation of all rotor connections

5.check for correct wiring BLK/WH/GRN to tube stator from J4

6.check for correct lead orientation to CR4 on AMX1A4 panel

7.check that R6 above autotransformer on AMX1A4(older units) isn't shorting

8.check for correct wiring of plug to J8 Fil/KVP board

9.check for square wave at C6 on AMX1A4 Panel

10.bad rotor

11.open rotor leads

12.bad ribbon cable to CPU from Rotor board

13.check lead tightness with AMX1A3C2 and AMX1A3R3

Contact me so we can work on this to resolution

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)